Could you try following parameter ?
Storage ==> CSV // DataSrc ==> MEM // Param ==> ON and sample depth > 60Kpts
Yep, same as you
its failed to save ![Face Palm :palm:](https://www.eevblog.com/forum/Smileys/default/facepalm.gif)
i already feed back this problem to Rigol, they said it will be fixed in next firmware version.
I received my DS1074Z-S today, with the firmware version 00.02.00.SP1
I have emailed Rigol about firmware update since this one has some bugs. Where am I suppose to request firmware by the way? (I just emailed their European sales office, it was the only email I could find)
- I can not enter the advanced system info menu either with this firmware.
- I can't save waveforms or CSV with large number of points, limited to very small memory, gives error otherwise.
- If I turn the scope off with the signal generator turned on, it will boot up with it turned on as well (remembers settings), but the status LED will not be on (signal is enabled but no indication that it is on).
- All advanced math functions (integrate, abs, sqrt etc.) only work when 2 channels are enabled (even though they should only need one channels), seems that you need to have specific combination of channels enabled: Ch1, Ch3 or Ch2, Ch4. Other combinations of active channels results in error message.
- In menu: Utility -> System, there is option "VerticalExp", with no explanation in the manual or in the help menu in the device. Can be set to "Ground" or "Center", which determines the reference when you zoom in on a vertically offset signal. Ground is standard, Center means that the entire trace shifts vertical position when you change vertical gain.
This was a few things I found during the first 2 hours of testing.