Hi,
I have a RIGOL DS1102E that has served me reasonably well for a year or so. Today, while visualizing a simple waveform (PWM output from a microcontroller through a series of resistors, 3.9KHz, ratio between 30 and 60%, low level of around 800mV and a high level of of around 3.4V), I ran across something I would not expect: changing the vertical position of the waveform on the screen changes the waveform when I'm in 200mV/DIV resolution or lower. The waveform distorts and shifts, so much that when the vertical position is at its upper limit, +2.0V, the signal appears with a lowest point of -1.7V (which is simply impossible). The plateaus also show a slope that depends on the vertical position (much more pronounced when moving the position up than when moving it down...)
Higher resolutions (500mV/DIV and higher) do not show the unexpected behavior, and moving the position does not affect the waveform, it stays a nice square between its expected value. Just like I would expect from a scope.
As a control, I tested again with the calibration test signal of the scope, with the same behaviour on all ranges (works OK on 500mV and more, not on 200mV and below).
DC coupled, Y-T, triggered auto, original probe 1x, CH1, nothing on CH2 and it's turned off. No math operation, no BW limit, no digital filter, no cursors, no measures, nothing fancy at all. Ground lead is well connected to GND.
Firmware version 00.04.00
Any idea what is happening? Is there an option somewhere that I could have activated that would make the scope act like this? Is it possible the scope is damaged?
Is it normal and have I accumulated too many all-nighters to understand what is going on (I'm doubting my own sanity at this point)?
Thanks