I have been just watching your review and stumbled upon your post.
You seem to be getting a few things wrong there regarding the hardware. The fuse ratings are written on them. Just rotate and you'll find their interrupt rating for both AC and DC written on them.
The R20 is a 1k ohm (not 1M ohm) fuse-able resistor. If MOVs started to conduct this resistor will open like a fuse and stop the fault current. Its a safety feature that you'll not find in cheap meters including most Uni-Trend meters.
The C1 10nF 1000V capacitor is for AC coupling. And R1 1K resistor is to discharge it when you rotate the function switch to some other function.
I am surprised about your comment about Uni-Trend UT61E+ current measuring capability that it compensates for the shunts temperature. I couldn't find any mention of it from your or other people's reviews or in its manual. Would you please explain a bit more about it?
There is however a safety feature built-in that lights up a yellow LED, sounds the buzzer and shows "CUT" on display if the internal temperature rises to or above 75 degree C while measuring current as per the manual.
Fuse ratings are not written on them. There are just some fractions, and god knows what they mean. Well, They actually mean current, but industry standard, at least here in Europe is to write the amperes straight. Like "440mA" or "0,44A". There is "DMM-44/100-R" written on it. This sounds just like a generic part code, but not fuse rating. What is the purpose of using fractions in such cases where you use SI units? We don't write 20mA as 2/100. Such a stupid idea it is to express amperage by fractions.
The R20, well, I must apologize for I got that wrong. I somehow remembered the 1 meg value and then additional 10 meg. This confusion may come from other multimeter I have look into. So thanks for clarification, will mention it in the desc.
So, the C1 is for AC, okay, thanks for info.
UT61E+ is the only multimeter with classic shunt wire, that even when overheated, still has the same reading, no matter the shunt temp. I have overloaded the multimeter (1 minute at nominal current should be enough for CUT indication), and the measurement was still absolutely correct (after the CUT was done beeping). Any other multimeters begin showing more and more, until you get like 12A on display while just 10A is passing through the multimeter. Usually shunt wire gets hot at nominal current, begins to expand its length and increases resistance, so also increasing burden voltage, causing bad reading. Some SMD shunts are prone to this too, but not that significantly of course.
Regarding the CUT function, there is one thing to it. It doesn't engage the warning when you are on different range and you keep passing current through your multimeters A -> COM. This may happen only to some newbies though.