Nice to see you TomC again
By the way, my wishes for your birthday (soon)...to be very well with your family.
I am looking about fw, with yours serial and the result was the same Ver. 3.3!
You have downloading and installed the fw to your scope or downloading only and saw some txt file?
You have fine memory, mine was 2012 series (1211xxx).
Sorry Lemon, I didn't catch what you were trying to tell me with the pic when I first read your post, yes, that's the firmware I downloaded, and even though it says V3.3, it's the same file that is used for the later models listed as V3.4. That V3.3 is what I installed on my scope, but it gave me version 3.8, with apparently all the new features that the change log lists for the older newer models. Don't know why Owon makes the listing so confusing, it appears that the same file is used for the whole SDS line.
Exactly, the fw 3.3 for our series upgrade to 3.8 like fw 3.4 for the newers serial Numbers.
What a strange fw politic!
Sometimes, they improved the fw have the same number, but change the data of uploading file
Owon all the times are confuse us!
Just, I installed the new fw and here are my notes:
1. XY mode - persistence function added- I don't look it
2. waveform refresh rate improved- Seems to be some improvements cause changes at Autoset.
For that reason I repeat the test of Mark Madel (Owon SDS7102 review. youtube) with the trigger output capturing at Chanel II.
The Mark has found that this DSO varies the refresh rate depending time base & memory depth. He had recorded 22.96fps at 1MHz sinus signal (1K depth memory, 1us time base) that dropped to 16.43 when the memory depth was 10M.
With the same conditions, I repeat the test and the results seem to be very faster. With 1K depth memory the refresh rate was 119.07 and with 10M was 36.95. Nice to be good!
Please look at my attachment to verify.
3.waveform cutting/ recalling function added- Now, at
Save menu adding a type of "
CutWave", the message that appearing now is "
move cursors and press Save to save wave to Udisk. Recall wave btw two cursors by OWON generators".
In fact with this method saves data to .ota files that probably plays to Owon generators !!!
4. autoset - waveform recognition- Here is the big differences.
Now when you pressed the
Autoset button, the scope try to identify the waveform and if it is success then presents choices depends from what waveform is.
For sinus waveform there is continuous period sinewave or full one period and FFT.
For pulse or square waveform there is continuous pulses, one period, rise pulse, fall pulse
Triangle or something else doesn't recognized and no more choices are available.
Each time you set different waveform must to press the
Autoset button to recognized it.
It is pitty that the FFT choice doesn't work with waveform together (split window)
Have in mind, that the
Autoset works only if the
Autoscale is
OFF5. channel invert bug fixed- OK, it corrected.
6. probe attenuation compensation;- Another automatic procedure for relaxing. Connect your probe to a stable 1KHz/5V and scopes makes a diagnosis if this probe has a good compensation or not.
7. recorded waveform exported to USB drive- It seems to adding some sub-menus to record menu now.
8. automatic measurement - 4 more options available- They added week rms, cursor rms, duty cycle and Phase?RP
Still is missing the all measurements to real time, just each time that you pressed the "Show All" button.
9. vertical position over-speed movement bug fixed- It corrected.
Finally, there are some improvements with the more major to me the more stable function of Autoset.
Still missing basic functions like fine V/div adjustment, split window for zoom or fft and something else.