I've replaced the board but had faced with another issue: there ia a large interferences caused by backlight LED DC-DC converter. Earlier on old board I've discovered some weak interference from backlight (
https://www.eevblog.com/forum/product-reviews-photos-and-discussion/review-of-owon-sds7102/645/), but it disappeared when I connected a 50 Ohm BNC terminator or probe, even when I connected a 10kOhm resistor or any signal source with any output impedance there was no interference, so it did not affected measurements because it was visible only when nothing is connected to inputs.
I've compared peak level of noise of my oscilloscope with screenshots which other people shared on the Internet and found it is about the same. Thus I've considered that it is OK, now I have reasons to gues that it is not OK
.
With new mainboard interference is much larger and they are enlarged when I connect terminator to input. I've installed the old mainboard again and see the same very small interference visible only on open input. Thus it is not due to incorrect installation - because I swapped boards few times and always I get negligible interference with old board and large interference with the new one.
The problem appears when powered from battery or from AC mains. So it is definitelly not a PSU problem.
I've made screenshots with new boards. X10 probe configuration is selected for both channels. I just forgot to switch it to X1, thus V/Div readings on the screen should be divided by 10 to get the actual input sensitivity.
An interference on both channels is the same. I leave CH1 open (no probe connected) and connect 50 Ohm BNC terminator to CH2 to show that 50 Ohm terminator increases an interference.
NewBoard_1.png: CH1 is open, CH2 shorted to ground through 50 Ohm BNC terminator, brightness is 0%.
There RF noise extremely low for 2mV/Div sensitivity. Seems that it is even much lower than a noise on the old board.
But there is a terrible 30kHz interference from backlight LED driver.
NewBoard_2.png: CH1 is open, CH2 shorted to ground through 50 Ohm BNC terminator, brightness is 100%.
30kHz noise is much lower comparing to an interference with 0% brightness, but it is still present and spikes occasionally appears.
NewBoard_3.png: FFT for CH2 which is shorted to ground through 50 Ohm BNC terminator, brightness is 0%.
30kHz interference with harmonics is clearly visible.
NewBoard_4A.png, NewBoard_4B.png: Even with low sensitivity 30kHz interference is clearly visible. With 2V/Div interference is higher than with 1V/Div. Thus interference comes rather through power line than through input.
It doesnt look like a mainboard issue, because there are a lot of DC-DC converters on the adapter board and on the mainboard itself and only an LCD backlight step-up causes interference. I guess that a mainboard of new revision is just more sensetive to low-frequency ripple on supply voltages rails and a real problem is a LCD backlight step-up on the adapter board or new mainboard has some compatibility issues with old adapter board (I'm not sure that OWON changed adapter board, this is only an assumption). There are a lot of changes in analog and power management parts on the new board comparing to my old one.
Thus I would like to ask people who have SDS7102 - have you any signs of 30kHz noise with visible low display brightness with open input (no probes connected)?
Because I have this noise on two mainboards of different revisions - on old 1209*** it is wery weak and produces no visible impact on measurements, but it is still visible when nothing is connected to input BNC, especially on FFT. On new board (1211****) interference level raises when I connect a probe. I've recorded a video, you can see that on 2V/Div with an input is connected to probe compansation output an interference just makes any precission measurements impossible (while on 1V/Div it is much lower):
And the second strange thing - I've noted some strange behavior of external synchronization input when it is configured to trigger on falling signal edge, on old mainboard the trigger works fine, so owners of 1211**** devices, please check check is this issue appears on your devices. On the video both ext trigger input and a channel input are connected to probe compensation output and an AC coupling is selected for trigger and channel input itself: