Hi Sarasir,
I'm amazed at the number of mods you have made, have you devised a way of quantifying the improvement after each mod? Personally, I'm always curious about the effort/improvement ratio. Not that it matters much if the goal is near perfection, and it seems to me that you are well on your way toward achieving that!
As far as the noise tests, I'll try to explain what we have been doing so far and why. It all starts with the symptoms of what we've been calling GND noise in the SDS7102s. On scopes that are not otherwise defective, it manifests itself when the probe, set to X10, is used in combination with the 6" ground lead, or strap as you call it. On these scope the base line noise is negligible, about 1mV, and signals, including low level ones, are virtually noise free when the spring ground clip is used instead of the 6" ground lead. However, when the 6" ground lead is used there is HF noise with spikes riding on the signal, this noise persists even if both the probe tip and the 6" ground lead are connected to the scope's ground lug. The amount of noise varied quite a bit from unit to unit, in my case, it was in the range of 100-150mVpp, with the highest spectrum amplitude at around 100MHz.
Similar symptoms have been well documented when signals on a DUT are polluted with common mode noise. Scope manufacturers warn in their literature about this effect and often provide probe accessories to allow shorter ground connections, for example, the spring ground clip that comes with the Owon probes. Common mode noise is normally invisible to the scope, which is designed to view differential mode signals. However, because the 6" ground lead causes an impedance imbalance due to its inductance, the common mode noise is converted to differential mode noise, which is then naturally displayed by the scope.
The disturbing fact in this case, is that the common mode noise originates within the Owon, and even when viewing signals from a clean DUT, the user is deprived from the convenience of using the 6" ground lead. Member rf-loop, which is also an Owon dealer, was the first one to identify the SMPS and peripheral DC/DC converters as the culprits. To help members determine if their scopes had this problem and to what degree, a methodology accessible to most users was devised to check/test the ground noise. This basically involved using the scope probe set to X10, and connecting both the tip and the 6" ground lead to the scope's ground lug. The scope was then set to 50mV/div, 100us/div, maximum sampling rate, and several other details, before taking a reading. See the following post for all the details:
https://www.eevblog.com/forum/testgear/review-of-owon-sds7102/msg255469/#msg255469Several approaches toward minimizing this problem emerged from a number of our members with varying degrees of success. My approach was to try and mitigate the noise at the source by altering the circuit so it wouldn't cause so much noise. For my purpose, I was satisfied with the results. But many other members obtained similar results using other approaches, and Owon obtained similar results with the new PSU and a mod to the adapter board. None of these provided a perfect solution, but the noise was reduced to the point that it was possible to view low level signals while using the 6" ground lead.
I attached a few PDFs that were particularly helpful to me. If you are interested, next I can post images of some of the test results, like before and after noise levels.