Oh, and once I wanted to buy this Owon scope... Anyway, is this problem even at the 200MHz version SDS8202??
This whole terrible issue is in some manufacture lots.
Owon make some modification for main PSU (there is 2 SMPS) and adapter board what have also SMPS circuits.
Some schoolboy "EE" propbably do this work becouse of course this kind of small changes are easy.. only move some components and small changes in layout and then some components away what looks like do nothing... ... oh yes.. simple woirk and then becouse it is so simple directly order new PCB's for factory product line.
Yes, if look with signal generator etc.. all looks ok. So, go on...
But in test lab... they do not as usually some peoples do... connect probe GND to somewhere and center to somewhere.. professional peoples have mesurement practices and probe with GND longwire is only for some "blinking led" measurments.
Also I do not this if can avoid. Only for slow high level signals and so on in some service situation where only fast need look like "are there anybody home". And no problem with this common mode noise.
Also there was no any reason to measure GND. Why I measure and watch just GND with oscilloscope. Is it fun, or just becouse bored and nothing better to do..
But then, some really may use oscilloscope probe with long GND wire for low level signals or fast signals.
If there is noise, first thing is take this short GND spring.
Oh well... then connect probe longwire inductor as GND to scope GND together with probe tip... and voila...screenfull of noise... what hell is going on - nearly drop from my seat.
After then start investigations that what hell is going on and why...screen is full of noise.
If connect clean signal to input BNC all is ok and more than ok... it is very good ok, also if use normally used short GND there is no problem at all.
There is common mode high frequancy noise what is just converted to differential mode with this probing.
Now we know. Now also Owon know.
Later after I have time I will show some test pictures what ahave made using with one original "noisy GND" issue oscilloscope from factory and then just with exactly same tests using factory new unit after this issue is repaired by Owon.