Hi all,
Happy owner of 2 Keithley multimeters which I bought recently. 1 is a 2000 which is 100% functional, no self test errors. The second one is a 2001 which has substantial damage done by leaky caps. Somebody already did a repair attempt to it, but its definitely not 100%.
After some investigation this is where its at at this moment:
It does power on, let me take measurements, seems on the low voltages ranges its pretty accurate. Not really possible for me to check its accuracy anyway. It does jump up in voltage for no apparent reason once every few seconds for a very short time. Ohms range is not working well. Shorting the probes gives good readings, but anything above a few ohms sometimes give me an ok reading, but most of the time its al over the place.
Its missing its mains transformer. Probably it got a shorted turn, and got tossed away at some point in the past. Now it is running from an external transformer which is putting out the following voltages:
9.3v and 35.0v to analog board, and 9.5v to digital board.
I noticed that J1025 is not used atm. This is for the voltage selection. Do you guys think that this can give issues? Maybe the circuit is trying to sense and switch all the time causing issues?
I started to take some measurements but when I saw that the voltage at one of the avalanche diodes was at around 48v I stopped and thought it would be better to clean and replace all obvious problems first.
Failing self tests:
200.1 No A/D communication, noisy A/D
200.6 U808 Malfunctioning
200.7 U808 Malfunctioning
201.2 Cannot measure 7v at A/D
201.3 Cannot measure 1.75v at A/D
300.2 A/D Mux 1.75V Ref, x5 gain
304.6 70nA Ohms source
304.7 5nA source
What I found inside:
Analog board:
A/D converter board, running rev A06
Power supply caps replaced (not by me). Notice the bodge wire running under it. If you look closely you can already see the horror that happened between those 2 caps.
Burnt hole in PCB, and some surgery already done (not by me)
![](http://)
Bodgewire running from U510 all the way to U320
Corrosion around c108
Wierd pin on U626, original or not?
![](http://)
Half eaten away resistors near C633
C623 and c624 leaky
![](http://)
So lots of rework needs to be done. First of al I'm planning to change all caps, also the ones that already have been replaced. Just so all caps are fresh, and the same brand/type (I ordered Panasonic 105c ). I also want to put in the original transformer. Quoted Tek via parts request on the website, but no response so far (about a week now). Called them, and they said it is available for 220 euro shipped to the Netherlands (where I live). If somebody is willing to sell one for less than that, I'm happy to hear.
Lastly, I want to thank TiN, and definitely all the other who contributed to this magnificent work!!!
Thanks,
Tom
***Having trouble getting the pics online, will fix that tonight***