Hello!
I save a 2001 from trash in my customer. There was a label on the power plus with the mention 'don't power on before repair' and the fuse is missing.
Last calibration was on 2019, I would like to repair it to replace my poor DMM.
Then I open the device and see two things: the power transformer has suffering and there is two dead diodes.
PCB are clean, this units looks like there wasn't very used.
The dead diodes (CR514 and CR112) are relative to bootstrap voltage generation. I don't understand why CR514 was dead with CR512, schematic sugest if CR514 dead then CR511 or CR509 suffer also but not CR512 ?
Anyway the issue is relative to BS+ and BS- generator, these voltage rail seems to be used in input buffer section only. How can I test if there is issue in input buffer section ? Impedance between BS+ and BS- ?
The electrolitic capacitors looks really good but I read this is a recurrent faillure in these device.
Impedance of winding of transformer looks good (I find a picture on the net with impedance of each winding), despite the fact the transformer has been hot.
What repair plan can you suggest me ? In first approach I think about changing dead diodes and also relative electrolitic capacitor. I read all electrolitic capacitor should be changed for a good rework, maybe will do that.
And CR512 is a melf BAV103 component which can be found, but CR112 is a BYD17G which seems hard to found, anywone have a sugestion of alternate buyable part ? I think change all diodes relative to BS generator is good idea.
Thanks for any help anyone could provide