Seems K2000 use same CADDOCK assembly marked TF-255.
Ok, somebody was asking earlier what one should do for attempt repair 2001.
Here's starting point, as all seven units I got here had this issue more or less.
Due to age (my units were manufactured in 1992-1994), even good electrolytic
capacitors become bad and leak electrolyte on PCB. This can cause major destruction.
Problem is not always obvious, contrary to dodgy caps repairs like in PSU, as
these caps don't blow, but silently leak electrolyte under them.
Not a problem you say, PCB covered with protective soldering mask? Wrong!
Check it out (all photos clickable):
![](http://dev.xdevs.com/projects/kei2001/repository/raw/photos/K7/small/K7_fe_damage.jpg)
![](http://dev.xdevs.com/projects/kei2001/repository/raw/photos/K7/small/K7_555_damage.jpg)
That's utter damage done to traces, as it got under mask and eated traces, vias and pads.
Older revision L analog board elevate this problem even more, as near C114,C115,C116,C117 there are diodes and two zeners,
which fail that curcuitry, causing PCB burns.
![](http://dev.xdevs.com/projects/kei2001/repository/raw/photos/K7/small/K7_mux_damage.jpg)
![](http://dev.xdevs.com/projects/kei2001/repository/raw/photos/K7/small/K7_pcb_damage.jpg)
And electrolyte spreads around board, screwing everything around, which is really bad as analog board have all these high impedance signals around
which really sensetive to leakage and coupling. Cleaning this PCB is not an easy job. Not clean board often fails such self-tests like 304.6, 304.7 and 20MOhm,200MOhm,1GOhm ranges,
which require to source proper current levels in nanoAmp range.
![](http://dev.xdevs.com/projects/kei2001/repository/raw/photos/K7/small/K7_cap_damage.jpg)
This is how bad it can be
![Smiley :)](https://www.eevblog.com/forum/Smileys/default/smiley.gif)
And notice, its not some cheesy Samwa/noname capacitor, it's well-respected Nichicon VZ cap.
So as a story to learn, if you have 20-year old test equipment, not only Keithley 2001, take it apart and replace all electrolyte caps just to prevent such horrors happening. Capacitors, unlike voltage references, are not getting better with age.
As a bonus today:
Frequency test on passing 2001 #4
Speedup 400% on video, to make it short
Was able to measure almost 25MHz, even while spec only states up to 15MHz.