Hello All, not much progress has happened since 2018 except lately I do have now a TDS540C, a PG506 and found the P5210 manual on internet.
As opposed to my attempt to calibrate and/or repair the 2nd P5210 discussed in this thread backin 2018, the difference is that now I have connected per the instructions of the P5210 manual a 100V then 10V pulse generator at 1KHz. This time, I'm able thanks to the blue potentiometer mentioned by @Hydron have the same gain one each P5210. One P5210 was calibrated prior purchase in USA so it serves say as an etalon so now both CH1 and CH3 of the TDS540C receives each P5210 outputs (no need for the power supply tek 1103), each high voltage inputs of P5210 being connected to the single Standard Amplitude output fo my PG506.
As you can see now on the attached scope shots, wether 1000x mode or 100x each probe are within the tolerance at 1KHz.
However I notice on the partial un-calibrated P5210 there is an overshoot peak so I wonder first what part of the circuit could fix plus if this effect could not explain why at 250KHz sinusoidal I still cannot adjust the gain so both probes would roughly converge.
The key thing which is hard unless one of you have this information, the attached P5210 manual only explains what test to verify the integrity but does not have a specific service manual section explaining how to calibrate. I did try earlier connecting both P5210 probes to the 230Vac grid, the blue pot would sync each gains to be the same. However still the same problem if probing a 250KHz sinewave (I've used a 200W induction lamp ballast and bulb), namely pushing to its limit the blue pot there is no way to align same gain output.
Thank you, Albert