Surprised to see SMPS in this LNA to be honest.
It is massively shielded from radiated emissions but a single pass through a ferrite is all the exiting wires get. There does not appear to be any protection on the main PCB's to deal with over voltage either, so I am not encouraged. I cannot find a block diagram or schematic and there is not even a silkscreen to help.
With that said - I powered it up externally based on the voltages printed on the PSU. The logic comes up just fine in the internal precision voltage generator works. The inputs, however, appear to be dead. The internal voltage source appears at the output, but no signals appear to pass. The current draw is right at the stated max current of the power supply:
Stated capacity vs Actual
+29v
300mA 300mA
-29v
300mA 260mA
+5v
1000mA 850mA
And the analog section with all the discrete transistors get hot - certainly hotter than my other one. I checked the various OP AMPS to make sure the supply voltages were as expected, one of the voltage references checked out ok. The 5v logic seems to be normally operating - that rail was just dead and I don't think it ever saw any high voltage. The only rail that went high voltage was the +29v. The negative rail is the one pulling more current.
This is not looking good. A lot of mystery and not much time to chase it down. The eBay seller has not yet responded to my notes. I got so far as to explain that the PSU was spewing out random high voltage and likely destroyed or at minimum seriously compromised the rest of the amplifier. This thing is rather high-end and well outside of my understanding. I believe it is past the point of practical repair - at least for me. I am sure someone with a ton of patience could go from transistor to transistor looking for bad things. Even if I could find something and get it working - I would have to send it to LeCroy or someone for calibration adding to the cost. If it failed calibration, that would suck even worse.