OK, so I've been working on documenting my test setup for the verification of the design specifications.
I've already placed some test results of these tests here, and plan to release all orderly on Github.
It's a total of 12 measurements with their own setup. I've the text here, with pictures of the setup, or a HTML collated.
I'm open to discuss the test setups, let me know if you think I'm doing something wrong or you know of a better way.
Text is in markdown, the HTML is nicer, but not supported as attachment, so I have it as PDF.
Code: [Select]
# Measurements done on the PRP1 Power rail probe
This documents lists all the measurements that are done on the PRP1 for design characterisation.\
The measurements were done on TBD (at least two) separate units, results are published in folders
## 1) Bode plot 10MHz
Verifies the gain errors across the frequency span\
![image](1.png)\

Equipment used:
* Siglent SDS804X HD Oscilloscope
* Siglent SDG1032X HD Generator

Settings:
* 10Hz to 10MHz
* 121 point
* 100mV peak to peak
* 0V offset
* Frequency mode: Linear
* DUT input CH1
* DUT output CH3
* Load 50 Ohm

Pass criteria: 1dB maximum deviation between minimum and maximum of the gain plot
Note: Crossover frequency is ~ 100 KHz\
Note: 1.3:1 is 0.77 or -2dB

## 2) AC gain @ 10 MHz
Verifies the 1.3:1 probe attenuation\
![image](246.png)\

Equipment used:
* Siglent SDS804X HD Oscilloscope
* Siglent SDG1032X HD Generator

Settings on generator:
* 10MHz sine wave
* 100mV peak to peak
* 0V offset
* Load 50 Ohm

Pass criteria: 5% maximum deviation between RMS and Peak to peak value of CH1 and CH3\

## 3) Noise measurement 20MHz BW 12 bit Scope
Verifies the noise figure of the PRP1\
![image](31112.png)\
Settings:
* PRP1 zeroed-out
* PRP1 powered by lab linear power supply
* PRP1 input shorted by SMA shoring plug or 50Ohm load
* 2 mV/div on CH1 and CH3
* 1 ms/div time scale
* 20MHz bandwidth limit

Pass criteria:
$\ Noise = sqrt{CH3^2 - CH1^2) < 50uV$

## 4) DC gain
Verifies the 1.3:1 probe attenuation on DC\
![image](246.png)\
Equipment used:
* Siglent SDS804X HD Oscilloscope
* Siglent SDG1032X HD Generator

Settings:
* PRP1 zeroed out on Oscilloscope with 0V input
* 0.1 Hz square wave
* 100mV peak to peak
* 0V offset

Pass criteria:
Measured 100mV peak to peak on the oscilloscope on CH1 and CH3\
Two measurements are 5% within each other

## 5) Cancellation range +24V and -24V
Verifies the +/-24V Cancellation range of the PRP1\
![image](5.png)\

Equipment used:
* Siglent SDS804X HD Oscilloscope
* Rigol DP832 Power supply

Settings:
* PRP1 zeroed-out after each change of the voltage on power supply
* Set up 5V, 12V, 24V with 0.1A current limit on power supply
* Flip BNC to banana adapter and repeat measurement

Pass criteria:
PRP1 can be adjusted to less than 100mV measured on the oscilloscope
Output current of the power supply is less than 1mA

## 6) 500mV dynamic range
Verifies the active signal range of the PRP1\
![image](246.png)\
Equipment used:
* Siglent SDS804X HD Oscilloscope
* Siglent SDG1032X HD Generator

Settings:
* PRP1 zeroed out on Oscilloscope with 0V input
* 0.1 Hz square wave
* 500mV peak to peak
* 0V offset

Pass criteria:
Measured 500mV peak to peak on the oscilloscope

## 7) DC input 50KOhm
Verifies the 50KOhm input impedance in DC conditions of the PRP1\
![image](7.png)\

Equipment used:
* Siglent SDS804X HD Oscilloscope
* HP 34401A multimeter

Settings:
* PRP1 zeroed out on Oscilloscope with 0V input
* resistance measurement auto range

Pass criteria:
50Kohm +/- 2% measured on DMM

## 8)a VNA bandwidth 10MHz-3 GHz
Verifies the minimum 2 GHz bandwidth of the PRP1\
![image](8a910.png)\
Equipment used:
* LiteVNA
* Rigol DP832 Power supply
* MegiQ VNA calibration kit
* LiteVNA SS405 Coax cables

Settings:
* 10 MHz to 3 GHz sweep

Pass criteria: 1dB maximum deviation between minimum and maximum of S12 up to 2 GHz\
Nominal gain is -2dB +/- 1dB\
Note: 1.3:1 is 0.77 or -2dB

## 8)b VNA bandwidth 400MHz-3 GHz 2nd setup
Verifies the minimum 2 GHz bandwidth of the PRP1 with a more accurate VNA\
![image](8b.png)\

Note: Minimum frequency setting of the VNA is 400 MHz\
Equipment used:
* MegiQ VNA-0440e
* Rigol DP832 Power supply
* MegiQ VNA calibration kit
* LiteVNA SS405 Coax cables

Settings:
* 400 MHz to 3 GHz sweep
* S11,S12 measured

Pass criteria: 1dB maximum deviation between minimum and maximum of S12 up to 2 GHz\
Nominal gain is -2dB +/- 1dB

## 9) S parameters
![image](8a910.png)\

Equipment used:
* LiteVNA
* Rigol DP832 Power supply
* MegiQ VNA calibration kit
* LiteVNA SS405 Coax cables
Settings:
* 10 MHz to 3 GHz sweep

Pass criteria: None
S parameters provided for reference


## 10) 50 Ohm loading at high freqency, VSWR
![image](8a910.png)\
Equipment used:
* MegiQ VNA-0440e
* Rigol DP832 Power supply
* MegiQ VNA calibration kit
* LiteVNA SS405 Coax cables

Pass criteria: VSWR < 1.5 up to 2 GHz\
This is equivalent to Return loss < -13dB\
Or capacitive loading of ~ 0.5pF

## 11) Noise measurement 1GHz BW 8 bit scope
Verifies the high freqency noise content of the PRP1\
![image](31112.png)\
2 GHz scope should be used for ideal conditions, but I don't have access to one\
Equipment used:
* Keysight MSOX 3104T Oscilloscope
* Keithley 2221A Power supply
Settings:
* PRP1 zeroed-out
* PRP1 powered by lab linear power supply
* PRP1 input shorted by SMA shoring plug or 50Ohm load
* 2 mV/div on CH1 and CH3
* 1 ms/div time scale
* 50 Ohm input on scope

Pass criteria:
TBD

## 12) Power consumption
![image](31112.png)\
Equipment used:
* Siglent SDS804X HD Oscilloscope
* Rigol DP832 Power supply

Settings:
* 5V, 0.1A
* PRP1 zeroed-out

Pass criteria:
Power consumption less than 20mA

Managed to write frequency wrong 3 different ways.  :palm:
« Last Edit: September 20, 2024, 12:33:40 pm by tszaboo »
 

Online tszabooTopic starter

  • Super Contributor
  • ***
  • Posts: 7903
  • Country: nl
  • Current job: ATEX product design
Re: PRP1 - Low cost 2GHz power rail probe
« Reply #51 on: September 24, 2024, 11:03:20 am »
https://github.com/tszaboo/PRP1
I've placed the following on Github:
User manual
Design verification test results
Design verification updated guide

Important change that I increased the specification for the noise. I originally measured the noise with 1:1 setting and that's technically not correct, since it's a 1.3:1 probe, therefore the new specification is 65uV 20MHz.

For me, this concludes the design verification phase.
I'll continue to assemble the PRP1s, test them, and to order the accessory cables. One cable I also need to make myself.

Please see below an extra oscilloscope shot I made during the verification. It's noise coming from an USB power supply CH1 measuring it with regular oscilloscope probe, CH3 with the PRP1 with the 2.54mm to SMA cable. We can see how significantly less noisy the PRP1 is compared to a standard probe.
 
The following users thanked this post: Kean, _Wim_, 2N3055

Offline 2N3055

  • Super Contributor
  • ***
  • Posts: 7197
  • Country: hr
Re: PRP1 - Low cost 2GHz power rail probe
« Reply #52 on: September 24, 2024, 11:47:05 am »
https://github.com/tszaboo/PRP1
I've placed the following on Github:
User manual
Design verification test results
Design verification updated guide

Important change that I increased the specification for the noise. I originally measured the noise with 1:1 setting and that's technically not correct, since it's a 1.3:1 probe, therefore the new specification is 65uV 20MHz.

For me, this concludes the design verification phase.
I'll continue to assemble the PRP1s, test them, and to order the accessory cables. One cable I also need to make myself.

Please see below an extra oscilloscope shot I made during the verification. It's noise coming from an USB power supply CH1 measuring it with regular oscilloscope probe, CH3 with the PRP1 with the 2.54mm to SMA cable. We can see how significantly less noisy the PRP1 is compared to a standard probe.

Just wanted to say, nice work !!
Results look quite good.
 
The following users thanked this post: tszaboo

Online tszabooTopic starter

  • Super Contributor
  • ***
  • Posts: 7903
  • Country: nl
  • Current job: ATEX product design
Re: PRP1 - Low cost 2GHz power rail probe
« Reply #53 on: September 30, 2024, 08:38:09 am »
I had some time this weekend to test and put together the first batch.
I'm still waiting for the cables and a sample of the shipping box.
I expect I can call these ready in about 3 weeks time.
 
The following users thanked this post: Kean, KungFuJosh

Online Martin72

  • Super Contributor
  • ***
  • Posts: 6700
  • Country: de
  • Testfield Technician
Re: PRP1 - Low cost 2GHz power rail probe
« Reply #54 on: September 30, 2024, 09:28:39 pm »
What would it cost?
Since reading this article, I have become very interested in it:

https://www.siglenteu.com/power-integrity-measurement-solution/
"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)
Siglent SDS800X HD Deep Review
 
The following users thanked this post: KungFuJosh

Online tszabooTopic starter

  • Super Contributor
  • ***
  • Posts: 7903
  • Country: nl
  • Current job: ATEX product design
Re: PRP1 - Low cost 2GHz power rail probe
« Reply #55 on: October 01, 2024, 10:25:39 am »
What would it cost?
Since reading this article, I have become very interested in it:

https://www.siglenteu.com/power-integrity-measurement-solution/
250 EUR + Paypal and shipping fees.
Siglent is very right about those measurements. Switching noise can and will couple into passive probes. You are trying to measure ripple on a power supply rail, but those spikes make it impossible, since they superimpose and make a few millivolt ripple into hundreds of millivolts of noise spikes. Often times it's so short and it lines up with your trigger, you don't even see it on the scope screen. Plus, thats not what's on your board, it's something that couples into your 1MOhm/10MOhm passive probe as EMI.
Now, power rail probes that are made for your scope are great, and if you (your company) can afford one, buy that. It shows the DC component, tells you if the probe is in overload. Much better for reporting and design verification. And they have cable compensation, or DC calibration. But they are very expensive. The PRP1 won't show you the DC offset on the screen, or tell you if it's in overload. So you have to be aware of the limitations. On the other hand, you can see from the scope screengrabs that I posted, it's less noisy than a x10 probe, and doesn't show the EMI switching noise.
 

Online tszabooTopic starter

  • Super Contributor
  • ***
  • Posts: 7903
  • Country: nl
  • Current job: ATEX product design
Re: PRP1 - Low cost 2GHz power rail probe
« Reply #56 on: October 01, 2024, 02:56:07 pm »
I've finally had time to finish the remaining test, the second VNA measurements with a higher quality VNA (Megiq). I could do a sweep between 400MHz and 4GHz on 3600 points, so every MHz.
S11, SWR and others are just informative, this is not a pure 50 Ohm system since the source impedance is 0 Ohm and the load is 50 Ohm, and I have parts in series. But it's still tells a bit about the system performance, so I included it.
More important is the gain VS frequency. Nominal gain is 1.3:1 or -2dB. Normally scope manufacturers define the bandwidth of a probe or a scope at the -3dB point compared to the nominal gain. The -1 dB point is 2.7GHz for this probe, and the -3dB is over 4GHz, meaning that the probe fulfills the 2GHz design requirement. I doubt that there are many people with 2GHz+
 scopes that would be interested in buying this.
The choice of cabling and the scope's internal 50 Ohm (or external 50 Ohm) will be the limiting factor in all cases.
« Last Edit: October 01, 2024, 02:59:24 pm by tszaboo »
 

Online tszabooTopic starter

  • Super Contributor
  • ***
  • Posts: 7903
  • Country: nl
  • Current job: ATEX product design
Re: PRP1 - Low cost 2GHz power rail probe
« Reply #57 on: Today at 04:22:44 pm »
Cables arrived, packaging sample arrived, and it's a nice one.
I only have to order a bit more from the packaging and I'll be ready to ship out the first batch.
 