Another thing I've noticed is that X plate voltages are very different if I center the trace. Something like 84V vs 50V. Is this normal because of properties of the tube or something fishy is going on in X amplifier?
Try to measure the x plate voltages in X-Y with spot in the middle of the screen. Be careful, not too much brightness as this can burn the screen.
I bought my new Hameg HM605 in 1982 and I have used it for 2 decades in tropical climates.
However, I have had many failures, but relatively easy to repair.
NB: this scope has thousands of hours of working !
Here's the list:
- CA3127 replaced twice in the same channel
- A high voltage multiplier defective replaced
- Two electrolytics replaced (they really exploded)
- Two bridge rectifiers with interrupted diode. (very commun failure, better replace them by 4 diodes 1N4004)
- Various cold solders reworked
On my second HM605, I had to replace the TL820CN of the overscan circuit.
NB: for frequency precision measurement, I use the backpanel BNC Y output conected to a 5315A HP universal counter.