AC-current measures today, in the 4A range - Under 1A it´s not usable.
Measurings taken with appx 3.9A and various frequencies.
"Reference" was Fluke 187.
60Hz 400Hz 1khz 4khz 10khz 20khz
Fluke 3.86 3.87 3.84 3.75 3.54 3.14
LeCroy 3.84 3.82 3.79 3.67 3.35 2.77
CP05+ 3.89 3.90 3.81 3.81 3.64 3.21
So far so good…
But:
I´m sure I did a zero-offset before, both lines ( AP011 and CP05) were on zero-scaling before starting to measure.
Now I watch the screenshots at home…
(Ch1 Ap011, Ch 2 CP05+, for better viewing not in the same current flow direction)
Seems like a positive offset, althoug I did the zeroing before measure.
And:
The phase deviation between the two channels by increasing the frequency.