I have a new Siglent SDM3045X multimeter which is not measuring frequency correctly.
The test frequency being used is a 5 volt p-p 10kHz rectangular signal which has it’s pulse width adjusted to show that at certain pulse widths the SDM3045X meter does not read frequency correctly. On the meter I have tried using the auto as well as the 6v range and the rise time and fall times are 20nS.
Here are the test results:
Applied pulse width - SDM3045X frequency reading
1uS - correct
5uS - correct
10uS - correct
20uS - correct
28uS - no reading
revert to 20uS - no reading
30uS - random reading
31us - correct
52uS - random reading
54uS - no reading
82uS - correct
95uS - unstable
96uS - no reading
I have used other input frequencies with similar results. I have tried ac coupling as well as dc coupling the meter input, without an improvement.
It is mostly problematic as one can never be sure of a frequency reading without studying the symmetry of the wave. Sine wave measurements are correct.
On contacting the Siglent factory I got this response; “Yes, the measurement for the pulse signal will be not stable. Usually the frequency measurement is more suitable for the sine.”
Has anyone solved this issue?