The current drain on the battery is about 3mA. Battery life must have been very good.
The voltage multiplier circuit contains 2 2N1307 PNP Ge transistors for the oscillator. I can hear a very high pitch noise when the counter is on.
There is also 4 old style PH204 silicon diodes. They tested Ok on my multimeter. With 4 diodes, it's probably a voltage quadrupler at the output of the transformer.
The other board contains 3 2N3392 NPN Si transistors.
The PC boards are very rustic. No silk screen. No solder mask. Everything looks hand soldered.
There is no nameplate, no serial number anywhere.
The counter wasn't mass produced, for sure.
So, what we have is a circuit that produce a high voltage for the Geiger tube.
Every time a particle or gamma ray hits the tube, a small current impulse is generated. The current gets amplified, integrated and displayed on the meter.
With the audio output of the counter and the radioactive mantle, I could check the calibration. All I need to do is check if the pulse count match what the needle indicates.