Hello to all.
I've checked this thread from start to end.
I bought 2 units (HDS272s) from (bgd) CZ depo -
Both have v3.2.0 firmware version and SN are starting with 2128xxx .
Testing the units I found or confirmed several firmware bugs.
I might be wrong so maybe some of you will do same tests and confirm or not - "it's not a bug is a a feature" style.
------------------
a) "Channel 2 is Off" on screen without any reason when using HOR menu or the Trigger menu.
How to test: just press HOR or Trigger menu. The message will pop up.
(@
TheBay noted this bug and I can confirm it).
b) User interface not responding in Waveform Generator menu.
How to test:
Use Mode to switch to Waveform Generator - set some frequency using arrow keys.
Use Mode to switch to Oscilloscope - use any of Channel / HOR / Trigger menus.
Use Mode to switch back to Waveform Generator try to use arrow keys to change Frequency (Frequency will show in green color that means is selected).
c) Waveform Generator doesn't change selected signal type:
How to test:
Connect oscilloscope probe to Waveform Generator output.
Use Mode to switch to Waveform Generator. Set Waveform to "Arbitrary".
Use Mode to switch to oscilloscope and check waveform.
Use Auto + Trigger to get a stable waveform on display.
Use Mode to switch back to Waveform Generator and set Waveform to "Sine".
Use Mode to switch back to oscilloscope and check waveform.
d) Osciloscope measured frequency is flawed / wrong when Acqu Mode is set to "Peak detect"
How to test:
(Use Mode to switch to activate oscilloscope).
In HOR menu check that Acqu Mode is set to "Sample".
In Measure menu activate Frequency for the channel that you are using.
Connect oscilloscope probe to calibration port.
Check the measured frequncy on display ( it should be 1KHz).
Use HOR menu and set Acqu Mode is set to "Peak detect".
Check the measured frequency on display (on my units 2.2 Khz instead of 1KHz).
e) In Waveform generator, when High and Low Level is selected instead of Amplitude and Offset, incrementing or decrementing last digit (second after the comma) changes both values of High and Low instead of selected parameter.
In some cases Down arrow or Up arrow will not respond to keypress and not change the values of selected parameter.
(@
saresigi spoted this bug - and I can confirm it)
I have videos that presents the aboved mentioned bugs and a return/repair notice was sent to bgd on December 28.
------------------
For those that are interested about HDS2102S: 2 non-public videos are posted on a japanese blog. Just search based on probe model included with these HDS devices.