Author Topic: over current in clamp meter  (Read 5029 times)

0 Members and 1 Guest are viewing this topic.

Offline hussamaldeanTopic starter

  • Supporter
  • ****
  • Posts: 266
  • Country: iq
over current in clamp meter
« on: November 25, 2018, 04:04:09 am »
Hi all,
I am planning to buy clamp meter to measure the current for motors, high power led and so on. but the question is what happen if the current was beyond the limit of the clamp meter. let say a clamp meter with 60A and the current in circuit was 100A ?
will the meter get damaged or remains functional ?
I know it is stupid question but I want to know
regards
 

Offline 0culus

  • Super Contributor
  • ***
  • Posts: 3032
  • Country: us
  • Electronics, RF, and TEA Hobbyist
Re: over current in clamp meter
« Reply #1 on: November 25, 2018, 06:29:31 am »
I would think it wouldn't be a problem since there is no actual electrical contact between the meter and the conductor you are measuring current through (unlike current measurements with a traditional multimeter). Assuming, of course, something else doesn't blow up.  :-BROKE

Fluke has some resources with FAQs:

https://www.fluke.com/en-us/learn/best-practices/test-tools-basics/clamp-meters/what-is-a-clamp-meter

It looks to me like the primary consideration is choosing one that works in the ranges you are planning to measure.
 
The following users thanked this post: hussamaldean

Offline graybeard

  • Frequent Contributor
  • **
  • Posts: 442
  • Country: us
  • Consulting III-V RF/mixed signal/device engineer
    • Chris Grossman
Re: over current in clamp meter
« Reply #2 on: November 26, 2018, 03:35:22 am »
I would check the manual to see what limits the manufacturer specifies and stick to those limits.  Since they designed it and if it is a quality piece of gear it should specify the maximum overload limits.
You could blow a fuse in the meter since the clamp is a current transformer that causes a smaller current to flow in the meter, it would be an annoyance but not the end of the world.
If you grossly exceed the maximum current rating of the meter it is possible you could induce enough heat in the core to cause damage.  If the overload is of a thermally short duration it could be fine.
It is also possible the core would saturate at the higher current peaks and be fine.

Offline NiHaoMike

  • Super Contributor
  • ***
  • Posts: 9222
  • Country: us
  • "Don't turn it on - Take it apart!"
    • Facebook Page
Re: over current in clamp meter
« Reply #3 on: November 26, 2018, 05:11:29 am »
It might magnetize the core causing an offset until it's degaussed.
Cryptocurrency has taught me to love math and at the same time be baffled by it.

Cryptocurrency lesson 0: Altcoins and Bitcoin are not the same thing.
 

Offline graybeard

  • Frequent Contributor
  • **
  • Posts: 442
  • Country: us
  • Consulting III-V RF/mixed signal/device engineer
    • Chris Grossman
Re: over current in clamp meter
« Reply #4 on: November 26, 2018, 09:03:41 am »
possible, but unlikely with an AC signal

Online Fungus

  • Super Contributor
  • ***
  • Posts: 17184
  • Country: 00
Re: over current in clamp meter
« Reply #5 on: November 26, 2018, 09:11:01 am »
Hi all,
I am planning to buy clamp meter to measure the current for motors, high power led and so on. but the question is what happen if the current was beyond the limit of the clamp meter. let say a clamp meter with 60A and the current in circuit was 100A ?
will the meter get damaged or remains functional ?

It's just a magnetic field, no permanent damage can occur.

It might need degaussing afterwards if you went way over the limit though

(depending on the clamp and how much you went over).

 

Offline hussamaldeanTopic starter

  • Supporter
  • ****
  • Posts: 266
  • Country: iq
Re: over current in clamp meter
« Reply #6 on: November 26, 2018, 09:59:16 am »
Hi all,
I am planning to buy clamp meter to measure the current for motors, high power led and so on. but the question is what happen if the current was beyond the limit of the clamp meter. let say a clamp meter with 60A and the current in circuit was 100A ?
will the meter get damaged or remains functional ?

It's just a magnetic field, no permanent damage can occur.

It might need degaussing afterwards if you went way over the limit though

(depending on the clamp and how much you went over).
thx a lot it will be 600A DC-AC TRMS one
I won't reach more than 20A in my experiments and maybe some other stuff for home usage like heater
 

Offline Neomys Sapiens

  • Super Contributor
  • ***
  • Posts: 3268
  • Country: de
Re: over current in clamp meter
« Reply #7 on: November 26, 2018, 11:40:48 pm »
All of the above holds true for CLAMPMETERS only. When you use an external current clamp (multimeter accessory), which is supposed to have the multimeter operating in current mode, you have to take care not to leave it unconnected. Because in this case, you will have a large voltage on the clamp's secondary winding. The manufacturers warn about this mainly to avoid injury, but I assume that their open circuit voltage rating is only good for the nominal maximum current. In this case, open circuit + overcurrent might damage the secondary.
Pure Hall Sensor clamps (and clampmeters) should be safe in this regard, whereas compound devices (like the fast current clamps from Tek etc.) are affected, because they have a secondary winding.
 

Offline coppercone2

  • Super Contributor
  • ***
  • Posts: 10613
  • Country: us
  • $
Re: over current in clamp meter
« Reply #8 on: November 27, 2018, 12:28:24 am »
it sounds kinda like the clamp should have a circuit that interfaces with the MM to open a connection, or at least a enable button.
 

Offline joeqsmith

  • Super Contributor
  • ***
  • Posts: 11871
  • Country: us
Re: over current in clamp meter
« Reply #9 on: November 28, 2018, 03:44:50 am »
I bought a UNI-T UT-210E clamp meter.  One of the first things I did was crank up the welder.   A few strikes and the meter would over range with no current.   No damage.  Just a quick degauss.

I should mention that after that first test, there was enough offset that the Rel would not zero out the meter.
« Last Edit: November 28, 2018, 01:16:24 pm by joeqsmith »
 

Online HKJ

  • Super Contributor
  • ***
  • Posts: 3023
  • Country: dk
    • Tests
Re: over current in clamp meter
« Reply #10 on: November 28, 2018, 07:24:53 am »
I played a bit with magnetization on clamp meters here: https://lygte-info.dk/info/DMMClampDemag%20UK.html
 
The following users thanked this post: rsjsouza, 001, Neomys Sapiens, hussamaldean