Author Topic: Is the Fluke 115 suppose to do that? or its a fault?  (Read 796 times)

0 Members and 1 Guest are viewing this topic.

Offline swapnil.vengTopic starter

  • Newbie
  • Posts: 2
  • Country: in
Is the Fluke 115 suppose to do that? or its a fault?
« on: June 13, 2019, 02:07:15 pm »
Its about my Fluke 115 Multimeter. Its displaying the correct DC voltage when its set to mV AC, but when I set it to V AC, it displays 0000. Shouldn't it display 0000 when set to mV AC and when the signal is DC? Everything else works fine. Even an AC signal in mV works properly.
I put photographs for reference.
« Last Edit: June 13, 2019, 02:17:16 pm by swapnil.veng »
 

Offline Wytnucls

  • Super Contributor
  • ***
  • Posts: 3045
  • Country: be
Re: Is the Fluke 115 suppose to do that? or its a fault?
« Reply #1 on: June 13, 2019, 02:34:05 pm »
All normal. AC mV position is DC coupled as shown by the symbol over the mV script.
Read page 11 of the user manual.
« Last Edit: June 13, 2019, 02:38:23 pm by Wytnucls »
 
The following users thanked this post: swapnil.veng