Hello again:
We make both time and frequency tests with our analog and digital scopes. Reliable specs are important in both domains. We rely on Tektronix and Yokogawa for specified performance in both BW and transient response. We have see many Chinese knockoff scopes and probes that have poor results in either or both areas.
1/ Depending on the application, BW or transient resp may be important.
Examples: Transient: Testing of high speed digital signal transmission eg transformers and cables.
Bandwidth: Checking generator outputs, broadcast exciters and transmitters, etc.
Notice that a good BW scope and probe could have a poor transient response and a perfect transient response will not have the widest BWs due to phase response.
In CAL of wideband scopes eg TEK 2465/7/B or 7104 + 7A29, transient response is the most difficult and critical adjustment.
BW is easier and if transient is good, probably BW is OK after CAL.
2/ We also use direct connect with coax. Many wideband and RF PCBs in fine instruments have builtin tiny coax connectors or probe tip receptacles for testing, esp TEK and HP of 1970s...1990s vintage.
3/ A well calibrated probe and/or wideband scope should pass BOTH BW and transient tests:
0.5 db, 3 db and BW = 6db with 50 Ohm term
Transient rise/fall time, abberation, preshoot, overshoot, etc.
The pulse source must be verified for fidelity.
VSWR of both scope and pulse source will cause reflections and give a worse transient result.
Standard practice is to insert a 6 - 20 db wideband attenuator at the scope 50 Ohm input, to attenuate reflections and improve VSWR.
We highly recommend the Mini_Circuits BNC><BNC inline attenuators, HAT-xx+ series eg HAT-10+ 10 dB, HAT-20+ 20 db.
Specified to 2 GHz.
https://www.minicircuits.com/pdfs/HAT-20+.pdf $10 ea VSWR 1.05..1.10.
4/ tests that use cables are naturally dependent on the cable transient and BW performance. ALL CABLES ARE NOT CREATED EQUAL!
Ordinary coax ><bnc or F or RCA CANNOT be used for calibration or fine measurements! Even USA made Pomona are NOT precision Zo or BW. Let alone the Chinese clone junk calbes.
Special calibration coax are available from TEK and GenRad, eg the 0.5m precision 50 Ohm 012-0482-00 BNC><BNC or the Rg8U GenRad 874 hermaphrodit connector 1,2,5,10 nS cables
Best connection is direct eg a BNC F><BNCF adapter and NOT a cable!
5/ BNC terms eg 50 Ohm can have issues eg poor quality connectors and inaccurate resistors. The internal construction and type of R determines the BW. Best inline BNC><BNC feedthru are the TEK and Mini-Circuits. The cheapo terms cant be relied on.
Enjoy!
Jon