Calling Fluke meters cheap is not entirely correct. I just checked Fluke 179 and 289 and they both goes between 10 and 11Mohm.
Compared with high-end bench top DMMs with active input tracking amplifier, providing many GOhms of input impedance.
My complaint is about the variation in resistance and obtuse specifications detailing it rather than the magnitude as long as it is high enough. It is absurd (except to the marketing department) to have high resolution meters with a variation in input resistance which causes additional errors greater than the basic accuracy or the 2nd to the least significant digit between ranges. Too many times now I have either spent hours tracking down a phantom circuit problem (1) or helping someone else do so when it was the multimeter lying.
As far as Fluke, what am I to think about the Fluke 25, 27, 83, 85, 87, and 88? One Fluke reference (2) warns about the variable input resistance of these meters but the manuals unambiguously say 10 megohms "nominal" without indicating that the input resistance is between 10M and 11.1111M and likely only "nominally" 10 megohms on the most sensitive range. My multimeters which have an invariant input resistance of 10 megohms +/-0.1% are "nominally" 10 megohms. That is marketing at its finest desiring to eat its cake and have it to look at. What the user does not know will not hurt them, or at least will not hurt Fluke. At least some manufacturers indicate greater than 10 megohms or list exact values. (3)
The difference between 10 and 11 megohms may not seem like much; after all, they are both high, right? But what source resistance would create as much error as the multimeter's basic accuracy or even worse, 10 counts or more with a change of 1 megohm between ranges? For 1% it is about 100 kilohms. For 10 counts on a Fluke 87, it is about 5 kilohms!
So Fluke is cheap except of course in price. They are not offering anything that their cheap competitors are not offering except their name and often less.
(1) It is usually a non-linearity problem which appears to be caused by excessive input bias current or poor common mode rejection.
(2) The manual for the Fluke 80k-40 high voltage probe gives detailed instructions for compensating for an input resistance other than 10 megohms. So just plug-in the value for the range and get a correction factor. Hmm, so what value do I use for "10M nominal"? Is it the same "nominal" value for all Fluke meters?
(3) The Fluke 179 specifications say "> 10M" for all volt ranges. The Fluke 289 specifications say "10 M" for all volt ranges but is that "nominally" meaning it could actually be 11.1111M or something else? I do not know what to believe.