v127 for 703 is very bad
I went into scope with prob set 10x ch1, changed freq to 1kHz sine 1v, back to scope and hit AUTO, it combed a few time scales to 100ms, then the unit seemed to freeze/crash and screen shut off to black, but 703 is still powered on! I press power button and flat line ch1 appears, then some just seconds later screen goes black again (Auto Off setting is set to OFF, I don't see any sleep or screen timer). Screen goes dark every 30sec.
(sine wave with DC coupling seems to be the culprit for AUTO with "crash", switch it to AC coupling and AUTO seems to be "ok")
1v on 10x probe config should not be hard for 702/703 to display, but weird stuff.
Also, using square wave, no longer can set freq or v or duty cycle. It's now just 1kHz 50%.
Measuring 1v sine, is off. 1v set, vpp says 860.0mv, or, the actual setting of 1v for the signal is off, not sure which.
Depending on volt scale chosen, the vpp measure changes, anywhere from 850mv all the way to 1.5v !
Set freq out sine 5kHz 1v (arbitrary setting), hit AUTO, then go into menu and cycle the Trigger Mode from Auto to Manual to Single and back to Auto, on the last back to Auto, scope sticks in STOP mode.
Scope will not allow you to turn off both channels at the same time.
I try the Calibrate menu item (open ports), it displays "Under Calibration" and then 5sec later a "Check Failure" notice (assume to mean claibrate failed). Where is the fail notice? It did not write any user accessible file into any folder when in USB mode.
Something not right with v127.
Timescale issue at 100ns seems to be fixed, I can tune all the way down to 10ns with the square wave.
Note: the F1 + power on is the same things as turning on USB from scope menu.