I've recently acquired a 1054Z and I love it. I've been playing with it, messing with the functionality and trying to replicate bugs. I expect everything to have bugs, but I want to be sure I know what they look like and how to reproduce them so that I am able to recognize them should they occur when I am actually taking a measurement.
The RMS bug seems to be the major remaining issue since the last firmware update (which I have done) and I noticed something odd with it. I'm using the 4 probes that came with it on the pins of a 47ohm resistor and am feeding a 1kHz sine wave from a function gen (yeah, a cheap FY3200S). The Vpp is 4.8V, so the Vrms should be ~1.71V. I can reproduce the bug just fine, I get 1.69V, 1.81V, 1.82V, 1.83V for the RMS on CH 1-4 respectively (My EEVBlog meter reads 1.680V). Nothing new or interesting there, Ch1 is correct and the other 3 are high. I can disconnect the probe for channel N and N+1 drops down to 1.69V. Again, nothing new, exactly as reported by others. However, when I disable channel N with the probe connected the relay clicks and the trace disappears, however the value on CH N+1 doesn't drop! I can even physically disconnect the probe after the channel has been turned off and the "extra" is still appearing in CH N+1. If I turn CH N back on (without reconnecting the probe), CH N+1 then drops to ~1.69V (±0.02 or so).
This I think cements that it is truly a problem in software, and some value somehow bleeding over.
I can take screen shots, but I think this table would be clear enough, and more concise.
NB: All channels on, All probes connected, triggering off CH 1 and all values are the Rms measurements.
CH1 CH2 CH3 CH4
1.69V 1.81V 1.82V 1.82V
Disable Ch2
1.69V ***** 1.82V 1.82V
Disconnect Ch2 probe
Enable Ch2 probe (still disconnected)
1.69V 691mV 1.69V 1.82V
Reconnect Ch2 probe
Same as above, but performing the same cycle with Ch3
1.69V 1.81V ***** 1.82V
1.69V 1.81V 683mV 1.69V
Same as the first two, except I moved the trigger to Ch4
and am performing the cycle with Ch1
***** 1.81V 1.82V 1.82V
27.8mV 1.67V 1.82V 1.82V