Author Topic: Fluke 8842A getting rid of a small current offset (-30 uA)  (Read 2753 times)
Offline SharpEarsTopic starter
« on: September 20, 2015, 02:45:51 am »
Is it possible on the Fluke 8842A multimeter to get rid of a small current offset (say about 30 microamperes or so)? This appears when the unit is fully warmed up and in current measurement mode and there is nothing attached to it. I was looking through the manual and short of going through the entire calibration procedure for all functions, I see no way to get rid of a fairly small current offset (i.e., to zero out the reading when there is no current flowing through the instrument).
If anyone has any thoughts or knows of a way to do this, please help me out.
Offline TiN
  • Country: ua
    • xDevs.com
Re: Fluke 8842A getting rid of a small current offset (-30 uA)
« Reply #1 on: September 20, 2015, 03:09:58 am »
Just use math REL function to null readout?
Offline deadlylover
« Reply #2 on: September 20, 2015, 03:13:20 am »
Does pressing the offset button work?
Offline mtdoc
« Reply #3 on: September 20, 2015, 04:02:11 am »
Yep, that's exactly what the OFFSET button does.  See section 2-29 in the manual "Offset Measurements".
« Reply #4 on: September 20, 2015, 02:27:46 pm »
I must have been blind, the Offset button did the trick! I guess I was looking for a Zero button and tried to search for zero/zeroing in the manual, as it is called on my other DVMs. Apparently, the word offset didn't click in my head as "compensating for an existing offset."