Some more bugs to report (none are show stoppers):
- Probe de-skew doesn't apply to trigger point, only waveform. If triggered off a de-skewed channel then the trigger appears in the wrong place.
- Cursors move the wrong one (Y) in X cursor mode at first
- Cursors menu sometimes shows wrong channel selected
- FFT max hold corruption on boot (very minor, just looks bad when booting and FFT max hold is enabled)
- Ref waveform doesn't seem to line up exactly with trace it is copied from (vertically with spectrum, horizontally with time domain view)
- When going into quickmeas and changing acquisition mode for one channel, doesn't seem to always apply to other channels when exiting quickmeas
- Pattern gen output can get stuck high when turning off
- In history mode, I disabled channels 1, 3 and 4 to get a better view of channel 2 in a past acquisition, but when re-enabling the channels to look at other captures only channel 1 appeared again correctly, 3 and 4 were flat (see "history bug 2.png")
This list is just my notes I made at the time of finding each issue; Rich/Markus - please PM me if clarification is required for anything.
When the next firmware is released a comprehensive list of fixed and remaining known issues would be very useful to let users know what they should report and what is already known (first firmware release had a list but it was much shorter list than expected).