What are the conditions for the sample rate for Keysight 1000x, 2000x, 3000x if you want to set the cut-off frequency of a LP filter at 10 Hz?
Running in high res acquisition mode
3000x: 10Hz LP fills just over half the screen at 50ms/div and 2MS/s (or 1MS/s or 500kS/s)
1000x: 10Hz LP fills just over half the screen at 50ms/div and 500kS/s
If you enable segmented mode then it is possible to reduce the sample rate further at the same timescale and the displayed filter trace remains the same length. So those scopes will show the filter where it makes sense and enough time is available for the filter to display sensible readings (rather than some initial condition guess)
For the HD3 what is interesting is that there is a new parameter in the signal chain to the math channels:
ADC - decimate/filter -> acquisition memory - variable decimation (?) -> measurement memory
there it should be possible to have the user trade off sample rate (measurement accuracy) with performance, but that needs some people with a HD3 to play around and test how much difference is available. Will it still do the slow filters at higher sample rates? probably.