Author Topic: New Keithley DMM6500 and now DAQ6510  (Read 330881 times)
Online Hydron
Re: New Keithley DMM6500 and now DAQ6510
« Reply #1475 on: June 07, 2024, 04:46:43 pm »
I wonder if the better screen is an off-the-shelf part that can be retrofitted?
0x2102 - if you're willing to open up the "good" unit to the point where you can see the details of the panel then that might shed some light on it? You can read some writing on the FFC on mine (poor viewing angle panel) with just the lid off, but I'm not sure that includes the required info - will do some googling.
Offline Caliaxy
« Reply #1476 on: June 08, 2024, 12:31:13 am »
I can definitely confirm that there are two different type of displays for the DMM6500.
The top unit has a calibration date of 3/2023 (display not as crisp, viewing angle limited) and the bottom unit 5/2023 (display sharp, no issues with viewing angle).
Unfair comparison. The upper display still has the lousy protecting film on it, the lower one doesn’t seem to have it. I would be shocked if removing it doesn’t improve the viewing experience :)
Online Pinörkel
« Reply #1477 on: June 08, 2024, 09:21:03 am »
As I had of course tested on my device, the protective film has no significant influence on the poor contrast or the screen doubling. It only adds environmental reflections. You can clearly see that in 0x2102's photos, where the settings menu area between the buttons is nearly black on the lower display(as it would be on the upper one when viewed from straight on) and medium gray in the upper one. This does not happen to the 100% black area around the voltage readout value, which clearly hints to a display tech contrast issue.
Offline Caliaxy
« Reply #1478 on: June 08, 2024, 02:04:55 pm »
I see, thank you… Also, have you tried adjusting the brightness setting (page 2-6 in the manual) to bring the dark background closer to black?
Online Pinörkel
« Reply #1479 on: June 09, 2024, 01:16:28 am »
Since this only regulates the backlight brightness, it naturally cannot mitigate angle dependent loss of contrast or screen doubling. The strength of the issue scales with backlight brightness, so it remains at the same relative strength to the black value of the display.
Offline Caliaxy
« Reply #1480 on: June 10, 2024, 12:41:27 am »
Well, on my DMM6550, adjusting the brightness seems to change a bit the contrast as well (black stays black at both extreme brightness settings, greys get darker at one end and brighter at the other end) - that's why I asked. Not sure in what extent it changes the contrast while viewing from an extreme angle though, if this is the issue the bothers you.
Haven't noticed any doubling...
Online Hydron
« Reply #1481 on: June 10, 2024, 09:48:02 am »
If you cannot see any doubling then you may already have an improved screen. This was a known issue (acknowledged by Keithley) at certain vertical viewing angles and is fairly obvious.
If you're game to take the lid off then you may be able to confirm by taking a pic of the display FFC where it peaks above the analogue board.
Offline 0x2102
« Reply #1482 on: June 10, 2024, 10:50:30 am »
I cannot see any doubling on any of my screens. The biggest difference between them:
1.) Contrast
2.) Viewing angle (one is worse when looking from above / other angle seemed to be fine)
3.) White balance (one is warm (bit yellow'ish and the other is colder (blue)
The one I prefer has a bit better contrast, viewing angle good from all side with a slight warm white balance. Part# ATM0500D27B1-CTN3 (seems to be an IPS panel)
The other one is Part# ATM0500D19F-CTN3 (TN panel)
« Last Edit: June 10, 2024, 11:15:36 am by 0x2102 »
The following users thanked this post: thm_w, Hydron, Pinörkel
Offline ArdWar
  • Country: sc
« Reply #1483 on: June 10, 2024, 11:15:12 am »
Of course D27 is much nicer, it's an IPS panel. D19 on the other hand is TN panel.
« Last Edit: June 10, 2024, 11:27:47 am by ArdWar »
The following users thanked this post: exe, Hydron, Pinörkel
Online Hydron
« Reply #1484 on: June 10, 2024, 07:03:48 pm »
Thanks for the info, and it appears that the text on the FFC is sadly not relevant shows the D19 vs D27 type. Mine is D19  :-BROKE
Looks like digikey has the nice one, assuming the trailing N3 in the part number isn't relevant. They clearly can't ship to the UK though, you can only find in on the US site, and it's a bit spendy at 70 bucks.
« Last Edit: June 10, 2024, 07:27:35 pm by Hydron »
Online Pinörkel
« Reply #1485 on: June 10, 2024, 07:05:08 pm »
Indeed, this explains the issues: TN-panel and viewing angle down 50°.
http://www.zettlercn.com/products/displays/tft-modules/atm0500d19-ctn-detail-id_m00_164.html
http://www.zettlercn.com/products/displays/tft-modules/atm0500d27-ct-detail-id_m01_186.html
Unfortunately there does not seem to be an easy source for the touch variant atm0500d27-ct. Most sources like digikey only seem to have the non-touch variant.
« Last Edit: June 10, 2024, 11:02:21 pm by Pinörkel »
Online Hydron
« Reply #1486 on: June 10, 2024, 07:06:32 pm »
https://www.digikey.com/en/products/detail/az-displays/atm0500d27-ct/13236094
I assume .de is blocked access to the part like .co.uk is
Edit: damn, need the B-CT variant which is not stocked (600 vs 350cd/m2)
« Last Edit: June 10, 2024, 07:11:54 pm by Hydron »
Online Hydron
« Reply #1487 on: June 11, 2024, 04:24:28 pm »
Quick update - I just checked another DMM with a factory cal date of 22/6/2023, and it has a marking of "FPC0500D27B1" on the FPC leading to the display assembly, i.e. it also has the better display fitted.
It's hard to tell how much better this display is as the meter is in an extremely bright room, but it doesn't seem to show the "doubling" seen on the older part.
I'm gonna have a poke around digikey etc looking for easier-to-source compatible displays (need to get a compatible touch controller as well as a bright IPS panel), will post again if I find anything promising.
Edit: swapped the panels and the old meter is happy driving the new panel (I assume the reverse is true, will test tomorrow). Also noticed that while digikey doesn't have the D27B-CT SKU, they do have the D27K-CT, which is a panel rated for a higher LED current (and proportionately higher brightness). If the LED drive is a simple current limited supply then this might be the easiest drop-in replacement, with the caveat that we don't know if D27B1-CTN3 (what's on the sticker in 0x2102's post) is the same as D27(B/K)-CT (what comes up on the web/digikey). 73 bucks though! :-\ https://www.digikey.com/en/products/detail/az-displays/ATM0500D27K-CT/12168275
« Last Edit: June 11, 2024, 08:39:15 pm by Hydron »
The following users thanked this post: jimjam, exe, 0x2102
Offline 0x2102
« Reply #1488 on: June 11, 2024, 08:48:31 pm »
@Hydron: maybe something to watch out for is that D19/D27B's background color is white vs the D27K's background color is black.
Online Hydron
« Reply #1489 on: June 11, 2024, 09:07:04 pm »
Slightly confused here - are you talking about "Display mode", which is normally black on all D27 SKUs and normally white on D19? That's expected due to TN vs IPS. Or is there a difference within the D27 options that I missed?
Offline 0x2102
« Reply #1490 on: June 11, 2024, 10:04:11 pm »
Never mind, I was slightly confused ;-)
Yeah, I was referring to the "display mode" but that makes sense TN vs IPS.
Online Hydron
« Reply #1491 on: June 12, 2024, 01:17:45 pm »
OK so just confirmed that the D19F-CT3 (TN) and D27B1-CTN3 (IPS) are cross compatible both ways between older and newer DMMs (at least when you clip in the touchscreen controller FFC properly... :palm:)
I think I'll leave further investigations to others, though if I do end up grabbing a screen next time I'm in the states I'll report back.
Offline Hexley
« Reply #1492 on: July 02, 2024, 12:20:21 am »
There is a firmware update available for the DMM6500. No new features, but a bug fix.
The version is 1.7.14j.
Details at https://www.tek.com/en/support/software/firmware/model-dmm6500-firmware-revision-1714-and-release-notes .
The following users thanked this post: HighVoltage, skander36, eplpwr, JimKnopf
Offline JimKnopf
« Reply #1493 on: August 08, 2024, 08:56:01 pm »
I need some help with Diode Mode Measurement.
I got a Crossfit-Clock here for repair. It has six large (122 x 90 mm) 7-Segment LED. Four red color LED (CPS40011DR8) and two blue color LED (CPS40011DSB).
During testing in Diode Mode, i found the red LED elements in working condition.
My default setting for Diode Mode is 10V,  1mA. The red element lights up when testing.
The blue LED elements looks like dead. No color, no reading, no sign of life. Because i couldn't find blue LED as spare part, i ordered red ones.
After desoldering the blue elements, i tought, what, if i use some higher voltage from Lab-PSU for testing. I tested with 0,01A, increasing at 7V. To my surprise, after reaching 10V, the LED came to life and showed up in blue color.
I then played with the Diode Mode settings.
The blue LED seems in working condition. But The DMM6500 has a weird behavior at 1mA in Diode Mode.
At 100µA, the LED light was more bright than at 1mA. Of course, at 10mA it was a strong blue color. I didn't expect that a DMM setting could lure me into a wrong judgment of my faulty DUT.
I attach some pictures with my test settings. I added a shadow for better view. The last two pictures represent my default light situation. As you can see, at 1mA, the blue color is invisible. I would probably have been able to recognize the blue color at 100µA.
Any idea why that happens at 1mA? Why does it works at 100µA?
« Last Edit: August 08, 2024, 09:56:00 pm by JimKnopf »
Offline Kleinstein
« Reply #1494 on: August 08, 2024, 09:16:12 pm »
The current source setting can effect the open circuit voltage. For the normal ohms mode the open circuit voltage with 10 mA and 1 mA is not that relevant as resistors in the low ranges are measured with a voltage of 1 V or less. Higher voltage would be too much power to be accurate.
The diode mode may use the 10 V voltage range, but this does not mean the current source could still deliver current up to that voltage.
Offline JimKnopf
« Reply #1495 on: August 09, 2024, 06:42:35 am »
@Kleinstein The LED is more bright with 100µA or 10mA.  Only 1mA has this issue. Looks like a bug. I assume, if the DMM6500 couldn't provide 10V/1mA in Diode mode, Keihtley would not provide this setting.
My device has the latest firmware 1.7.14j on it.
Offline bateau020
« Reply #1496 on: August 09, 2024, 08:14:05 am »
It is visible on the latest FW DMM6500: when the DMM says "overflow", don't expect full current.  I see no serious bug. Maybe you should measure Voltage and Current independently to get a feel of what it does in reality.
I do agree that it is not really homogeneous.
What I see on my latest FW DMM6500:
At 10mA and 1mA, the max voltage at full current is about 8V, but the voltage displayed already says "overflow" at 7.1V.
At 100uA, the max voltage at full current is about 12V, but the voltage displayed already says "overflow" at 10.0V
At 10uA, the max voltage at full current is about 12V, and the voltage displayed says "overflow" at 12V (and shows "X." instead of "10." or "11.", THAT may be a bug)
In all current ranges, except 1mA, the open circuit voltage is 12V. On 1mA, the open circuit voltage is 8.8V.
>> This is probably what you saw, but the DMM already told you that it would not be reliable.
« Last Edit: August 09, 2024, 08:26:45 am by bateau020 »
Offline pettel
« Reply #1497 on: August 10, 2024, 12:25:11 am »
I received my DMM6500 this week (Calibration 02/2024). The hum was very noticeable. I could alter the sound by pressing on the case. Besides the hum i'm really impressed by the instrument. It's really sad, that Keithley didn't solve this issue once and for all after so many years. I'll return it in exchange for another unit and hope for the best.
Online skander36
« Reply #1498 on: August 10, 2024, 10:51:06 am »
First try to change the operating voltage to 240 V - https://www.eevblog.com/forum/testgear/new-keithley-dmm6500/msg5488096/#msg5488096
You need to need to rotate the socket of the fuse. Extract and rotate. Is in manual.
The following users thanked this post: pettel