Characterizing an Advantest R3361B Spectrum Analyzer Input with miniVNA in the 1-180MHz range.
As per my knowledge of VNAs this should be a very easy REFLECTION measurement.
According to the results I get, ...it is not.
I know: the setup itself is questionable due to the differences between Golia (the R3361B) and David (the miniVNA). But this is what I have in hand and it is an experiment done for the purpose of learning something, not to make actual measurements.
Equipment:
• ADVANTEST R3361B Spectrum Analyser;
• vanilla miniVNA (no Pro, no lite, no v2, ...nothing);
• VNA-J software versione 3.4.8;
• BNC adaptors from China (!);
All calibrations are made for an OPEN over a 1-180MHz range with 6k points.
To make a prove of work I first calibrated the miniVNA for REFLECTION with a M2M BNC adapter followed by a F2F adapter. The calibration plan is expected to be located at the free side of that last adapter (far left end in the picture):
Connecting a 50 Ohms load ….
I’ve got exactly what I expected:
What is worth noting here is the constant behavior over the entire range (forgetting the spikes at the far ends of the range).
I then replaced the F2F adapter with a BNC2F_M adapter and re-calibrated the VNA:
then I connected the whole thing to the SA:
and here are the plots with different ATTENUATOR values:
SA input ATT = 0 dB
SA input ATT = 10dB
SA input ATT = 20dB
COMMENTS:
• The ATT= 0dB attenuator plot is ugly.
• The ATT=10db is a little better.
• The ATT= 20db is acceptable.
Why the first two plots are so ugly? • wrong setup?
• wrong vna?
• wrong SA frontend (input attenuator)?
• wrong data interpretation (what I see is correct);
• the experimenter doesn’t know what is doing and needs help and/or pointers to training resources...
Thanks.
Emanuele.