Author Topic: Accuracy of LCR meter?  (Read 4047 times)

0 Members and 1 Guest are viewing this topic.

Online Martin72

  • Super Contributor
  • ***
  • Posts: 6515
  • Country: de
  • Testfield Technician
Re: Accuracy of LCR meter?
« Reply #25 on: July 01, 2022, 09:53:52 pm »
Ergo it’s better not to worry about a couple of percent and a few counts and work out what the tolerance and device spread of your design is and make sure you window values within that.

Exactly.
You won´t design a coil into a circuit with a tolerance under 5% in most of the cases.
The AL value of the used core makes it impossible for mass production without cost intensive pre-selections.
So tolerances about 20% are not uncommon - Every LCR should handle this..
Of course there are some exceptions, e.g. coils for filters to suppress unwanted harmonics.
We (company I work for) make most of our used coils and transformers ourselves.
And test/adjust them.
We got several different LCRs and therefore different values of the same coil.  ;)
For the most the measuring with one of the LCRs are enough, for other we use resonant test circuits* for adjusting them.
Further we use Voltage-Time-Area function** or, with our self build tester, testing coils with large currents up to 200A***..
Based on this experiences, I can recommend LCRs building like the DE 5000 for most of the needs of a hobbyist.

Martin

*) Example: Filter-Coil, Air-Gap adjustment to reach given value 90µH +/- 3µH.
Using resonant circuit with additonal cap of 10µF/1%, adjust amplifier to 10Vrms, adjust airgap in such manner, that on 1.64....1.65khz the flowing current gets his minimum value.

**) Special case, self satured coils for zero switching, define windings on the coils per feeding in an impulse with high voltage, use math from scope, building integral function for voltage-time-area.

***) High power coils for output filters....normal LCRs didn´t have current, only a few mA max, our self build tester can deliver pulse-current up to 200A, adjustable.
The circuit based on a µC, some electrolytic caps for the current and a IGBT for switching.
L will be calculated based on the formular L= U*t/I (linear current is a must of course)

"Comparison is the end of happiness and the beginning of dissatisfaction."
(Kierkegaard)