Author Topic: New Agilent 34461A 6.5 Digit Bench Multimeter  (Read 171055 times)

0 Members and 2 Guests are viewing this topic.

Offline HighVoltage

  • Super Contributor
  • ***
  • Posts: 5547
  • Country: de
Re: New Agilent 34461A 6.5 Digit Bench Multimeter
« Reply #275 on: January 07, 2016, 10:12:08 am »
Yes, it was a software update, similar to the capacitance measurements, that was also updated with a new FW.

The dual display is not very well implemented.
For instance, if you measure DC Voltage and want to have the AC voltage as well on the second display it works with a delay, because it is switching a relay constantly between these two measurements.

This was much better implemented in the older 34410A and 34411A meters.

There are 3 kinds of people in this world, those who can count and those who can not.
 
The following users thanked this post: Andrew

Offline TiN

  • Super Contributor
  • ***
  • Posts: 4543
  • Country: ua
    • xDevs.com
Re: New Agilent 34461A 6.5 Digit Bench Multimeter
« Reply #276 on: January 07, 2016, 10:28:39 am »
TheSteve just broke my Keithley-heart, trying to xchange warm fuzzy VFD-DMM to that pesky cold evil LCD thingy, called 34461.  :popcorn:
YouTube | Metrology IRC Chat room | Let's share T&M documentation? Upload! No upload limits for firmwares, photos, files.
 
The following users thanked this post: Andrew

Offline HighVoltage

  • Super Contributor
  • ***
  • Posts: 5547
  • Country: de
Re: New Agilent 34461A 6.5 Digit Bench Multimeter
« Reply #277 on: January 07, 2016, 11:24:13 am »
I still have my many 34401A and 34410A meters in use and just like to use them more over the 34461A for most day to day measurement jobs.

But probably the best feature of the 34461A, and one that I am using all the time is the trend chart for almost all functions.
Even for small changes, the trend chart immediately shows it visually, before we see it in the numbers.

There are 3 kinds of people in this world, those who can count and those who can not.
 
The following users thanked this post: Andrew

Offline TheSteve

  • Supporter
  • ****
  • Posts: 3781
  • Country: ca
  • Living the Dream
Re: New Agilent 34461A 6.5 Digit Bench Multimeter
« Reply #278 on: January 07, 2016, 05:45:56 pm »
TheSteve just broke my Keithley-heart, trying to xchange warm fuzzy VFD-DMM to that pesky cold evil LCD thingy, called 34461.  :popcorn:

Ahh, so the TiN man has a Keithley heart - better then no heart at all! :)

I haven't made up my mind yet, shame you can't go somewhere and try these things out.
VE7FM
 
The following users thanked this post: Andrew

Offline Dr. Frank

  • Super Contributor
  • ***
  • Posts: 2425
  • Country: de
Re: New Agilent 34461A 6.5 Digit Bench Multimeter
« Reply #279 on: January 07, 2016, 06:44:13 pm »


I haven't made up my mind yet, shame you can't go somewhere and try these things out.

Shame on YOU, that you don't have a volt-nuts community in Canada?!

 
The following users thanked this post: Andrew

Offline sotos

  • Frequent Contributor
  • **
  • Posts: 257
  • Country: gr
Re: New Agilent 34461A 6.5 Digit Bench Multimeter
« Reply #280 on: January 08, 2016, 08:36:15 am »
ON mine, continuity is detected with no perceptible delay.  But the beeper, instead of being on continuously, emits a staccato beeping with variable period.  Is that the issue you're talking about?

You mean something like this.


 
The following users thanked this post: Andrew

Offline rosbuitre

  • Regular Contributor
  • *
  • Posts: 90
  • Country: ar
My instruments: DMM Keysight 34461A / Tektronix DMM916 / Fluke 12, Rigol DS1074Z, Deer DE-5000, Siglent SDG805 / SDP3303D, Dayton Dats2
 
The following users thanked this post: Andrew

Offline sotos

  • Frequent Contributor
  • **
  • Posts: 257
  • Country: gr
Re: New Agilent 34461A 6.5 Digit Bench Multimeter
« Reply #282 on: January 08, 2016, 11:13:44 am »
New version of firmware

http://www.keysight.com/main/software.jspx?ckey=2367633&lc=eng&cc=US&nid=-536902435.1058513&id=2367633

Regards

There’s nothing in the release notes mentioning the beeper problem fixed.
 
The following users thanked this post: Andrew

Offline neslekkim

  • Super Contributor
  • ***
  • Posts: 1305
  • Country: no
Re: New Agilent 34461A 6.5 Digit Bench Multimeter
« Reply #283 on: January 08, 2016, 03:49:24 pm »
And it's released 11th january?, here in Norway it's still the 8th..
 
The following users thanked this post: Andrew

Offline sotos

  • Frequent Contributor
  • **
  • Posts: 257
  • Country: gr
Re: New Agilent 34461A 6.5 Digit Bench Multimeter
« Reply #284 on: January 08, 2016, 09:53:49 pm »
And it's released 11th january?, here in Norway it's still the 8th..

Here too

 
The following users thanked this post: Andrew

Offline billfernandez

  • Regular Contributor
  • *
  • Posts: 141
  • Country: us
Re: New Agilent 34461A 6.5 Digit Bench Multimeter
« Reply #285 on: January 09, 2016, 04:24:35 am »
ON mine, continuity is detected with no perceptible delay.  But the beeper, instead of being on continuously, emits a staccato beeping with variable period.  Is that the issue you're talking about?

You mean something like this.




Yes, like that.   I find it disturbing, since I associate the cessation of the beep with a loss of continuity, which I have often found to be very useful.  This is just annoying, and makes me distrust the continuity sensing.
 
The following users thanked this post: Andrew

Offline HighVoltage

  • Super Contributor
  • ***
  • Posts: 5547
  • Country: de
Re: New Agilent 34461A 6.5 Digit Bench Multimeter
« Reply #286 on: January 09, 2016, 09:25:08 am »
Just don't use the 34461A for continuity measurements.

The good news: The 34470A does not have this problem.
The possible bad news: This could be hardware related and can not be fixed through FW updates.
There are 3 kinds of people in this world, those who can count and those who can not.
 
The following users thanked this post: Andrew

Offline sotos

  • Frequent Contributor
  • **
  • Posts: 257
  • Country: gr
Re: New Agilent 34461A 6.5 Digit Bench Multimeter
« Reply #287 on: January 09, 2016, 01:10:31 pm »
With Firmware, AGILENT V 1.10 its working fine, so I don’t think it’s a hardware problem. It gets messed up with Keysight updates.
 
The following users thanked this post: Andrew

Offline rsivan

  • Contributor
  • Posts: 43
  • Country: it
Re: New Agilent 34461A 6.5 Digit Bench Multimeter
« Reply #288 on: January 09, 2016, 08:15:45 pm »
Mine too have annoing beeper ,I did buy when still branded Agilent ,and with 1.10 Firmware was ok = "linear beep" not like now,this incredibile ,someone crazy like us making continuity test with professional bench dmm,just ordered Fluke 101 for 55€ shipped on ebay.de, and use for continuity test... very tiny and stay in  a pocket.
 
The following users thanked this post: Andrew

Offline TheSteve

  • Supporter
  • ****
  • Posts: 3781
  • Country: ca
  • Living the Dream
Re: New Agilent 34461A 6.5 Digit Bench Multimeter
« Reply #289 on: January 26, 2016, 06:54:38 am »
Has anyone had a go at finding the username/password for the 34461A telnet connection on port 5810?

Welcome to the Windows CE Telnet Service on K-34461A-14101

login:



edit: anyone have a link to download firmware 1.xx for the 34461A?
« Last Edit: January 26, 2016, 07:05:42 am by TheSteve »
VE7FM
 
The following users thanked this post: Andrew

Offline 6thimage

  • Regular Contributor
  • *
  • Posts: 181
  • Country: gb
Re: New Agilent 34461A 6.5 Digit Bench Multimeter
« Reply #290 on: January 26, 2016, 01:06:02 pm »
I have the 1.09, 1.10, 2.08 and 2.09 firmware files (NK.bin and build.xml - what the meter expects to find on the flash drive), but not the original update files. I'm not sure if you can get the old versions from the website (if so they will be on the firmware page), as I can't seem to get onto Keysight's website at the moment (503 errors).

I believe the username is 'Torreys', but I don't know of the password.
« Last Edit: January 26, 2016, 01:08:06 pm by 6thimage »
 
The following users thanked this post: Andrew

Offline plesa

  • Frequent Contributor
  • **
  • Posts: 965
  • Country: se
 
The following users thanked this post: Andrew

Offline TheSteve

  • Supporter
  • ****
  • Posts: 3781
  • Country: ca
  • Living the Dream
Re: New Agilent 34461A 6.5 Digit Bench Multimeter
« Reply #292 on: January 27, 2016, 09:36:03 am »
Awesome, thanks guys!
VE7FM
 
The following users thanked this post: Andrew

Offline billfernandez

  • Regular Contributor
  • *
  • Posts: 141
  • Country: us
Re: New Agilent 34461A 6.5 Digit Bench Multimeter
« Reply #293 on: January 27, 2016, 10:24:51 am »
Just don't use the 34461A for continuity measurements.

Yes of course.  Instead of using my new, $1,000 meter that's right there on the bench, the one that's capable of so many modern features executed with precision, I can drag out my old 1983-era Fluke 8060A DMM, which has awesome continuity response, to provide this most pedestrian of functions.  Or wait, don't I still have a beeper lying around from my high school days...?
 
The following users thanked this post: Andrew

Offline sotos

  • Frequent Contributor
  • **
  • Posts: 257
  • Country: gr
Re: New Agilent 34461A 6.5 Digit Bench Multimeter
« Reply #294 on: January 27, 2016, 12:54:41 pm »
Just don't use the 34461A for continuity measurements.

Yes of course.  Instead of using my new, $1,000 meter that's right there on the bench, the one that's capable of so many modern features executed with precision, I can drag out my old 1983-era Fluke 8060A DMM, which has awesome continuity response, to provide this most pedestrian of functions.  Or wait, don't I still have a beeper lying around from my high school days...?

You said it right.    :-+
 
The following users thanked this post: Andrew

Offline OldNeurons

  • Regular Contributor
  • *
  • Posts: 78
  • Country: fr
Re: New Agilent 34461A 6.5 Digit Bench Multimeter
« Reply #295 on: January 27, 2016, 04:50:25 pm »
1.10 and 1.09
http://www.filedropper.com/agt3446xareva0110
http://www.filedropper.com/agt3446xareva0109

Thanks for the files Plesa.

I downgraded my 34461A to v1.10 and confirm that continuity measurement works fine with this firmware.
So, for sure, this should not be a big deal for Keysight to fix that issue.
Does anyone here already reported to Keysight?
 
The following users thanked this post: Andrew

Offline sotos

  • Frequent Contributor
  • **
  • Posts: 257
  • Country: gr
Re: New Agilent 34461A 6.5 Digit Bench Multimeter
« Reply #296 on: January 27, 2016, 05:14:36 pm »
1.10 and 1.09
http://www.filedropper.com/agt3446xareva0110
http://www.filedropper.com/agt3446xareva0109

Thanks for the files Plesa.

I downgraded my 34461A to v1.10 and confirm that continuity measurement works fine with this firmware.
So, for sure, this should not be a big deal for Keysight to fix that issue.
Does anyone here already reported to Keysight?

If they monitor their site it’s here. I wrote about it and the link with the problem at YouTube

http://www.keysight.com/owc_discussions/thread.jspa?threadID=40301&tstart=0
 
The following users thanked this post: Andrew

Offline TheSteve

  • Supporter
  • ****
  • Posts: 3781
  • Country: ca
  • Living the Dream
Re: New Agilent 34461A 6.5 Digit Bench Multimeter
« Reply #297 on: January 27, 2016, 05:36:09 pm »
Yes, everyone who cares should reply in the thread on the Keysight forum. I will also be opened a case with Keysight directly, they need to put their big boy pants on and fix the problem.
VE7FM
 
The following users thanked this post: Andrew

Offline sotos

  • Frequent Contributor
  • **
  • Posts: 257
  • Country: gr
Re: New Agilent 34461A 6.5 Digit Bench Multimeter
« Reply #298 on: January 27, 2016, 05:58:40 pm »
Yes, everyone who cares should reply in the thread on the Keysight forum. I will also be opened a case with Keysight directly, they need to put their big boy pants on and fix the problem.

 :-+
 
The following users thanked this post: Andrew

Offline TheSteve

  • Supporter
  • ****
  • Posts: 3781
  • Country: ca
  • Living the Dream
Re: New Agilent 34461A 6.5 Digit Bench Multimeter
« Reply #299 on: February 17, 2016, 10:59:21 pm »
As posted by user "jdwindrvr" on the Keysight support forum today:

The "stutter" in the beep that signals continuity has been resolved. The firmware (with the fix) is going through our regression and production tests.

Assuming no unexpected issues will be found, the next release should be available in approximately four weeks.
VE7FM
 
The following users thanked this post: Andrew


Share me

Digg  Facebook  SlashDot  Delicious  Technorati  Twitter  Google  Yahoo
Smf