can i ask to measure the input impedance of the VC890 on each selector range w/o power on? and compare that to the fluke 101 / 87 and 1253B ?
A GDT for example is going to read an open. Of course, who puts GDTs in meters besides the Germans! Pulled out a few MOVs and the first one I looked at was 50M. It's going to be very non-linear and there is a lot to consider when it comes to protection than just the DC resistance.
The following data was taken with the Fluke 101 that beat out all the other meters in the transient tests. I waited for reading to settle, however settling time was several minutes in many cases. I may have not waited long enough in some cases but it should give you some idea. I used banana to banana leads. Sorry but I no longer have the old 87V to check. Besides it was damaged so may have different leakage after I changed out all the diodes in it.
No batteries, Pos lead to Pos, Neg to Neg ( I know you did not ask for polarity but it will make a difference)
ACV >100M
DCV 15.2M
ACmV 3.3M
Ohms 1.6M
Cap. 1.6M
Hz 3.1M
No batteries, Pos lead to Neg
ACV >100M
DCV 12.2M
ACmV 2.4M
Ohms 8.56K
Cap. 8.56K
Hz 3.1M
Batteries installed, Pos lead to Pos, Neg to Neg
ACV >100M
DCV 11.1M
ACmV 5.5M
Ohms 3.7 - 4.4M, Cont. 1.1M, Diode 4.8M
Cap. 1.8M
Hz 5.5M, DC 5.5M
Batteries installed, Pos lead to Neg
ACV >100M
DCV 11.1M
ACmV 5.1M
Ohms 2.3K, Cont. 2.3K, Diode 570K
Cap. 183K
Hz 5.0M, DC 5.0M
There is no manual range on the 101 to see how the attenuator effects it.