Sorry, i was out of counry on a business trip.
@retiredcaps: sure.
Continuity works, and i can measure a stable 1.628V between the input jack. It just doesn't display a ohms value
Temperature works as well at a stable 0.908V.
Capacitance doesnt work at all, when i put any capacitor across the probes, the meter switches to "----nF" and has 1.4mV across the probes.
Diode Tester does not work, but has stable 3.104V between the probes.
Edit: So far, i have only resoldered L3, all the other beads check out fine. R25 and R32 look a bit weird. i have not resolderd or reflowed those.
Now Ohms: Doesn't work, and the weird part is: Voltage across the probes starts at around 60mV, then goes up to about 350mV, then my cheap DMM (4000 Count i presume?) autoranges, and i get the reading of about 1,47V for a fraction of a second, before it starts at 60mV again. this whole cycle takes about a second.
Edit: A/mA DC range also not working. Shows O.L on both ranges with about 100mA going through the meter.