Very slow measurement from Rigol MC3065 DMM
« on: September 12, 2024, 08:10:56 pm »
Taking a voltage reading on this instrument is taking over 3 seconds which is an eternity. Haven't been able to figure out a way to speed it it up from the documentation. Anyone have any ideas?

Thanks.
 

Offline TurboTom

  • Super Contributor
  • ***
  • Posts: 1430
  • Country: de
Re: Very slow measurement from Rigol MC3065 DMM
« Reply #1 on: September 12, 2024, 09:00:52 pm »
You can set the integration time of the DMM module. Search the manual for "PLC" (Power Line Cycles) or see page 2-24 for further information on that topic. If this doesn't solve the problem, you may check if the measurement uses a trigger function to take readings at predetermined intervals. Setting up an M300 system can be a real PITA... Confusing to tell the least.  :-//
 

Offline stevesspTopic starter

  • Newbie
  • Posts: 2
  • Country: us
Re: Very slow measurement from Rigol MC3065 DMM
« Reply #2 on: September 12, 2024, 10:04:11 pm »
Thanks. Doesn't that only apply to DC though?
 

Offline TurboTom

  • Super Contributor
  • ***
  • Posts: 1430
  • Country: de
Re: Very slow measurement from Rigol MC3065 DMM
« Reply #3 on: September 13, 2024, 01:05:36 am »
I'm not sure. But since there's the RMS/DC converter contained in the DMM and thus the AC measurement is internally also a DC measurement, the integration time should apply here as well.
 