Author Topic: need help please  (Read 3940 times)
Offline defenderTopic starter
need help please
« on: June 15, 2014, 03:15:41 pm »
hello>
 how do i set oscilloscope for the following:
Set up a scope to monitor and sample the output
( from current probe ) with sample-time of 200 msec and sample-spacing
of 400 microsec
for a total of 500 counts     ( count=1 thru 500)
 thanks,
Offline Bored@Work
Re: need help please
« Reply #1 on: June 15, 2014, 03:18:04 pm »
When is that homework due? And do I get the credit points for the homework when I do it for you?
Offline FrankBuss
    • Frank Buss
« Reply #2 on: June 15, 2014, 04:04:34 pm »
I wonder what scope they are using where you can setup the sample-spacing individually, with a given sample-time, assuming you want one full screen width for the sample-time. And unless you are in some special sampling mode, sample-spacing is the same as 1/samplerate anyway, so why the fancy name which I had to Google? Doesn't sound like a practical homework where you can learn something for your work later.
Offline free_electron
    • SiliconValleyGarage
« Reply #3 on: June 15, 2014, 04:14:51 pm »
Without knowing what is the brand and model of the scope it is hard to say.
We're not going to write a manal for all possible scopes out there.
The question is bullshit anyway. Sample time 200 milliseconds (400 microseconds time 500 samples is 200 milliseconds.). Teacher is talking out of his ass confusing everybody. There is no need for this superfluous confusing information. It is ballast and can be construed as a trick question. Trick questions should be illegal in education. It's already hard enough conveying knowledge in a clear manner.
Sample time is for a screen, so total sweep time. There are 10 divisions on a screen. 200 milliseconds to traverse, at 10 divisions means : Set the timebase at 20millisecond per division.
If the screen is 500 points wide every point will be 400 microseconds apart.
Now, As for the 500 points.. That's gotta be the crappiest scope in the world. Even the very first digital scopes had 4k sample points...
« Last Edit: June 15, 2014, 04:20:48 pm by free_electron »
« Reply #4 on: June 15, 2014, 06:15:35 pm »
thanks for reply>
the oscilloscope used is  TEK TDS460 with TCP202 current probe.
here is the full test for for the electronic module i have to run.
 Set up a scope and a current sensing
device to monitor and sample the output
current pulses at J1-11 with sample-time
of 200 msec and sample-spacing of
400 microsec for a total of 500 counts
(COUNT = 1 thru 500).
Connect a 1 ? resistor between J1-11
and J1-12, and capture the first 200
msec of output waveform at J1-11
(COUNT = 1 thru 500). Capture the first
50 microsec of output waveform after
the pulsed DC output at J1-11 rises to
3.0 A (positive slope).
Remove the resistor between J1-11 and
J1-12.
For COUNT = 1 thru 500, and while
current measured at present COUNT is
greater than 3 A, do the following steps
with the initial value of TOTAL set at 0:
(1) Record the current
measured at present
COUNT as CUR
(2) Calculate TOTAL = TOTAL
+ (CUR * CUR)
Calculate TIME as follows:
TIME = (COUNT - 1) * 0.0004
Calculate TOTAL as follows:
TOTAL = TOTAL / (COUNT - 1)
Calculate  pulse ENERGY as follows:
ENERGY = TOTAL * TIME
thanks
Offline HighVoltage
« Reply #5 on: June 15, 2014, 06:22:17 pm »
Agilent has a pretty good educational powerpoint presentation for beginners.
Type this in google, it is the first link:
"Scope_Fundamentals_for_EE_Students"
May be this helps you understanding the basics.
Offline rf-loop
« Reply #6 on: June 15, 2014, 06:27:59 pm »
These are not at all very first digital scopes.
HP54501A, 54502A    501 sample points.
HP54200A/D  1001 sample points.
And for just same work what they have used in '80-'90 they are also today exactly as good as they have been new.  (but not for many today needs)
And today some 2GSa/s  scope may use example 28 points sampling buffer with 1ns/div setting.
What is crap - it depends...
