Thanks virtualparticles for the posts above!!!
My point about starting from a good known 50 ohms, even at DC, is better than starting a something other than 50 ohms if you intend to make lower frequency measurements as shown with these measurements.
Test setup:
NanoVNA SAA2N with SMA Cal kit (load DCR 50.931 ohms)
NanoVNA V2+4 with type N cal kit (load DCR 49.643 ohms)
Siglent SAA3021X Plus expanded to 3.2GHz and VNA enabled.
Another SMA Cal kit (load 49.687 ohms, short stubby type).
SMA test devices (SMA loads with long rear extensions) Green 50.882 ohms, Black 51.395 ohms.
All DCR measured with new KS34465A and fixture residual R nulled out.
SAA2N calibrated with supplied type N cal kit
V2+4 calibrated with supplied SMA cal kit
except when 49.687 ohm SMA used for calibrationSiglent SSA VNA calibrated with SAA2N type N cal kit.
Measurements at 10MHz utilizing averaging.
Short stubby SMA test load (DCR 49.687 ohms):
V2+4 48.7 ohms with RL of -37.92dB delta of -0.987 ohms
SAA2N 50.0 ohms with RL of -60.24dB delta of 0.313 ohms
SSA VNA 50.06 ohms with RL of -60.23dB delta of 0.373 ohms
SMA Green test load (DCR 50.882 ohms):
V2+4 49.9 ohms with RL of -54.04dB delta of -0.982
SAA2N 51.2 ohms with RL of -38.14dB delta of 0.318
SSA VNA 51.27 ohms with RL of -37.85dB delta of 0.388
V2+4 51.2 ohms with RL of -38.27dB delta of 0.318 (cal with 49.687 ohm SMA)SAM Black test load (DCR 51.395 ohms):
V2+4 50.4 ohms with RL of 46.51dB delta of -0.995
SAA2N 51.7 ohms with RL of -35.87dB delta of 0.305
SSA VNA 51.84 ohms with RL of -34.66dB delta of 0.455
V2+4 51.7 ohms with RL of -35.14dB delta of 0.305 (cal with 49.687 ohm SMA)A test at 100MHz with SMA test load (DCR 49.687 ohms):
V2+4 48.7 ohms with RL of 38.21dB delta of -0.987
SAA2N 50.0 ohms with RL of -41.69dB delta of 0.313
SSA VNA 50.19 ohms with RL of -41.03dB delta of 0.503
Note how the SAA2N and SSA VNA readings agree with each other and are more inline with the DUT load DCR value, much better than the V2+4 when calibrated with the supplied 51 ohm load. However the
V2+4 also agrees well with the SAA2N and SSA VNA when calibrated with a load that is nearer 50 ohms as shown in blue.
Note how the delta between the DUT DCR and the Cal DCR readings track. So it seems that starting from a reference near 50 ohms, the nearer the better, does indeed yield a superior measurement result at lower frequencies
Best,