So i am not sure if what i did is super correct but here it goes :
Test setup for noise measurement:
- Siglent scope SDS2104X
- 200 Mhz Bandwidth - this made sense to me since the current probe is 100 Mhz
- Pressed the 'ZERO' button on the probe
- Enabled the measurement function on the scope to show RMS, Mean, Peak to Peak and the Stdev measurement which i think but not 100% sure is the RMS without the DC component and the RMS contanints the DC component
- no cable through the loop.
This is what i can measure with the probe on the 6A range:
This is what i can measure on the 30A range:
Test Setup for Accuracy 6 A range :
- same as above + the following:
- DC electronic load
- Fluke 8846A Multimeter (not calibrated)
- DC power supply
Test Setup for Accuracy 30 A range :
same as 6A range
I am not sure how to interpret the
±10mA(6A) after the ±1%±10mA(6A) (spec from datasheet) maybe the experts can help here. I guess if i would take out the 10 mA somehow from the readings the probe could be in spec