Was it well calibrated when you got it?
If not then calibration can be initiated by doing (note you need a 6.5 digit meter as reference)
"shift" "up" "Down" "enter"
it then asks for a key:
enter "6666" and "enter"
display shows CAL
here are four types (that I know of) of calibrations:
readback V "V-set" button + connect a 6.5 digit voltmeter to the Vout terminals, M8811 generates the voltages, just enter what the external meter shows + "enter".
readback I "I-set" button + connect a 6.5 digit amp meter with 3A range to the Vout terminals, M8811 generates the currents, just enter what the external meter shows + "enter". (if calibrating a higher current version of the M88 series be sure to use an amp meter that can do near the full range setting, M8811 is rated at 5A and mine calibrated with 2.4A)
remote sense "Recall" button + connect a 6.5 digit voltmeter to the Vout terminals and connect the remote sense there too, M8811 generates the voltages, just enter what the external meter shows + "enter".
DVM "Save" button + connect a 6.5 digit voltmeter to the Vout terminals and to the DVM inputs, M8811 generates the voltages, just enter what the external meter shows + "enter".
power off and on again and calibration is done.
You can do the calibrations one by one.