Author Topic: SDM3055 Poor Autoranging and Wrong Measurements  (Read 1163 times)
Offline Cam765Topic starter
« on: February 24, 2022, 02:06:53 am »
Hello,
I've noticed before that my SDM3055 has had some difficulty adjusting it's autorange to the correct range but now it all together is off. It reads around 200 Ohms when nothing is connected and if I manually change the range the resistance changes to some value about halfway in the range. If I switch to 4 wire it drifts from anywhere between -100 and 100 Ohms (sometimes 0) depending on if I'm touching the front panel which is strange. I checked all the fuses and they're fine so if anyone has any helpful information that would be useful. I'll try to post some videos of it acting up if I can.
Thanks
Offline tautech
Re: SDM3055 Poor Autoranging and Wrong Measurements
« Reply #1 on: February 24, 2022, 02:34:31 am »
Some more history of this unit please.
How old is it ?
Have you the latest firmware version installed ?
https://int.siglent.com/download/firmwares/?ProId=26
« Reply #2 on: February 24, 2022, 03:03:31 am »
Sorry forgot to mention that it has the latest FW: 1.01.01.25
HW: 03-00-00-13-00
Device is less than a year old (about 9 months)
Offline tautech
« Reply #3 on: February 24, 2022, 03:21:36 am »
HW: 03-00-00-13-00
OK thanks.
I know of no other units that behave like this so I suggest you contact your supplier with the view to claiming warranty for a faulty unit.
Offline Kleinstein
« Reply #4 on: February 24, 2022, 08:28:42 am »
This looks like faulty unit.
The result in 4 wire mode is actually undefined if connection is made between the sense and drive part. Ideally the meter should give an error. In some cases it may auto range to the highest range and with some meters the high ranges are 2 wire only, even of 4 wire is selected.
Offline tautech
« Reply #5 on: February 24, 2022, 08:36:37 am »
It is.
Factory tech support requests it be returned to Siglent NA for repair.
Siglent NA have also been copied in so will be expecting it.
Cam765
Please contact: info@siglentna.com
Supply full info about the unit and a link to this thread is also a good idea.
The following users thanked this post: Cam765
« Reply #6 on: February 24, 2022, 09:47:58 pm »
Will do. Thank you all for the helpful info.