Author Topic: Marconi TK2374 200MHz zero loss probe for 2730. Can it be used with 2382?  (Read 576 times)

0 Members and 2 Guests are viewing this topic.

Offline brian_mkTopic starter

  • Regular Contributor
  • *
  • Posts: 88
  • Country: gb
The TK2374 200MHz zero loss probe was designed for the Marconi 110 MHz TF2370 analyser.
My question is: Is it compatable with the later Marconi 400 MHz 2382 (at least up to 200MHz)?

From what I've found so far, the Marconi 2388 1 GHz probe designed for the 2382 needs a +/- 15V power supply.
The TK2374 appears to use a floating 15v supply from the 2370 which it converts to -7.5v internally.

It's not clear to me if you can simply plug a TK2374 into the probe power socket of a 2382.

 

Online TimFox

  • Super Contributor
  • ***
  • Posts: 8593
  • Country: us
  • Retired, now restoring antique test equipment
I have successfully used the TK2374 probe with the 2382/2380 spectrum analyzer.
I was confused when trying to understand the power connections through the auxiliary cable, but it does work.
I measured the THD (at 500 kHz) of the combination, using a Wavetek 98 generator driving 0 dBm into a 50 ohm termination (either the SA input or a terminator at the probe input):

                    Direct          With 2374 probe
Fundamental      0                   0   
2nd harmonic   -63     0.07      -37   1.4
3rd harmonic   -61     0.09      -46   0.5
4th harmonic   -70     0.03     <-72   
      THD                   0.12                1.5
                  dBm      %            dBm    %

At 0 dBm/50 Ω = 0.22 V rms, the distortion of the probe is predominantly 2nd harmonic, with a total value of 1.5%.  The DDS generator itself is quite clean, despite being at one-half its maximum frequency.  The bandwidth of the probe (50 Hz to 200 MHz),  should not matter at this frequency.
 