Hi
Here I show how I do rise time measurements with Leo Badnar's pulse generator and from a Low Cost Function generator the Siglent SDG 1032X.
For good risetime measurements, it is important to use good cables and BNC connectors, together with an attenuator on the input of the scope.
My measurements are for scoops up to about 500MHZ and I have little or no experience with scoops above 1GHz.
The setup shown here is intended to minimize refelections in the measurement system.
Make sure the scope is set to 50 Ohm input impedance.
Use at least a 50 Ohm attenuator of -6 dB, in my measurement it is -20dB and preferably with a bandwidth of 2GHz or more.
I used a -20dB attenuator in this measurement setup, one from MiniCircuits and the other is from Radial.
The lower trace is from the Radial attenuator and the upper trace is from the MiniCircuits attenuator.
I cannot see any difference in attenuation between these attenuators, there is only a small difference in attenuation between the two types.
Only channel-1 was used, the attenuator on channel-2 is there to show you what types I used.
.
Both measurements with the two difference attenuators imaged.
.
But what does the block look like from an affordable feature generator....
Below the Siglent SDG1032X set up to reproduce the best block wave on the output.
This generator has a special circuit internally for that.
If you stay below 1V PP at 50 Ohm load, almost no abberations are visible.
The cable between the Siglent generator and the -20dB attenuator on the scope input is a BNC cable from Huber and Suhner.
.
And for completeness a picture when the Siglent generator is set above 1V, in this picture it is then 2V PP at a 50 Ohm load.
.
I hope this helps a little with showing rise times on a scope.
Kind regards
Bram