Hello,
I have a Metrahit Cal Calibrator which has a software problem:
When the unit switches off (or is manually switched off with the on/off button) while simulating Type K Thermocouple Temperatures it will switch to Type B Thermocouple when turning it on again. I have a quite old unit with firmware 1.15 and Gossen confirmed the software problem. But they can't tell me if it was fixed with a later firmware version.
So before I send the unit to Gossen for the firmware upgrade I would be very happy if someone, who has a Metrahit Cal with a firmware > 1.15, could check if that problem is still there.
Many thanks,
Magnus
Offline RobinLox
Re: Looking for someone with a Metrahit Cal to check something
« Reply #1 on: March 19, 2024, 04:43:00 pm »
Hello there, I just tried with a Metrahit CAL with Software 2.32 (which is the most recent one) and unfortunately the bug still persists.