Author Topic: looking for MASTECH MS8229 calibration procedure  (Read 13988 times)
Offline cyberfuzzyTopic starter
« on: August 08, 2015, 03:18:11 pm »
Hello,
I have several MASTECH MS8229 multimeters. But some of them are out of range. Unfortunately I can't find a calibration procedure for it. I asked the manufacturer for it but it seems so that they don't want to send me that information.
So I opened it up and tried to turn the trim pots. But I wasn't able find out what each potentiometer is for. There are some pictures of the pots on that page:
https://www.eevblog.com/forum/testgear/re-brand-of-mastech-ms2882()-from-harbor-freight-(item-98674)/
Has anyone of you a calibration procedure for the MASTECH MS8229? Or did one of you find out what the individual pots are for?
Thanks a lot!
Cyberfuzzy
Re: looking for MASTECH MS8229 calibration procedure
« Reply #2 on: August 08, 2015, 03:41:38 pm »
Thank you joeqsmith. I'm going to try to calibrate it now :-)
Offline retiredcaps
« Reply #3 on: August 08, 2015, 06:09:47 pm »
I'm curious about what is actually out range.  Is it DCV or resistance?  How much is it out of range by?
« Reply #4 on: August 09, 2015, 12:40:33 am »
@retiredcaps: I don't really know how much it is out of range because I do not have high end multimeters, voltage standards and so on. But to get an idea of it I did some measurements with four multimeters.
The number in front of the following results is the number of the multimeter as you can see here:
Multimeter 1: MASTECH MS8229 (original condition)
Multimeter 2: MASTECH MS8229 (original condition)
Multimeter 3: UNI-T UT58C (original condition)
Multimeter 4: MASTECH MS8229 (opened, trim pots turned)
Here are the results of some measurements:
DC Voltage:
(All multimeters connected to the same potential.)
1. 190.5 mV
2. 190.7 mV
3. 189.9 mV
4. 190.7 mV
1. 1.884 V
2. 1.889 V
3. 1.883 V
4. 1.894 V
1. 19.52 V
2. 19.61 V
3. 19.51 V
4. 19.70 V
1. 60.0 V
2. 60.3 V
3. 60.0 V
4. 60.4 V
AC Voltage:
1. 1.043 V
2. 1.043 V
3. 1.044 V
4. 1.037 V
1. 18.87 V
2. 18.90 V
3. 18.90 V
4. 18.87 V
1. 194.5 V
2. 195.1 V
3. 199.8 V
4. 194.0 V
1. 277.2 V
2. 278.1 V
3. 278 V
4. 276.6 V
Frequency:
1. 49.99 Hz
2. 49.99 Hz
3. 0.047 kHz
4. 49.99 Hz
1. 0.999 kHz
2. 0.999 kHz
3. 1.000 kHz
4. 0.999 kHz
1. 18.99 kHz
2. 18.99 kHz
3. 19.06 kHz
4. 18.99 kHz
DC Current:
(All multimeters connected in series.)
1. 35.91 mA
2. 36.20 mA
3. 36.0 mA
4. 36.37 mA
1. 188.2 mA
2. 189.3 mA
3. 188.6 mA
4. 190.1 mA
1. 2.912 A
2. 2.933 A
3. 2.94 A
4. 2.931 A
AC Current:
1. 0.402 A
2. 0.404 A
3. 0.41 A
4. 0.400 A
1. 3.024 A
2. 3.034 A
3. 3.06 A
4. 3.015 A
Resistance:
(I used a 1.82 kOhm, 1% resistor. LCR/ESR-Meter measured 1.8187 kOhm.)
1. 1.818 kOhm
2. 1.816 kOhm
3. 1.817 kOhm
4. 1.820 kOhm
« Last Edit: August 09, 2015, 12:43:48 am by cyberfuzzy »
Offline retiredcaps
« Reply #5 on: August 09, 2015, 12:53:36 am »
AC Voltage:
1. 1.043 V
2. 1.043 V
3. 1.044 V
4. 1.037 V
You don't necessarily need a voltage standard, but a quick look through your readings, I would say all 4 meters are within their specifications.  For the above measurements in your picture, the spec for ACV is 0.8% + 3 counts.  Assuming 1.043 is the "correct" measurement, all the meters are easily within the 0.8% +3 counts.
Even the best meters will differ 1mV or 2mV, but be within their respective specifications.
In your original post, I thought you were getting readings like 1.043 on one meter and 1.542 on another meter.
Offline retiredcaps
« Reply #6 on: August 09, 2015, 03:16:09 am »
DC Voltage:
1. 19.52 V
2. 19.61 V
3. 19.51 V
4. 19.70 V
The DCV for the Mastech is 0.7% + 2 counts.  The UT58C is 0.5% + 1 count.
If the 19.52V is the correct reading, then #4 is slightly out of spec.
19.52 x 0.7%  = 0.13664 = 0.14 => 0.14 + 2 counts = 16 counts.
So 19.52 + 0.16 = 19.68 (upper bound).
PS. I haven't done the calculations for your other ranges, but given the internal build quality and components of your meters, for most purposes, it should be good enough.  To get something more accurate, the meter has to have some high quality precision hybrid resistor network and/or software stored calibration offsets for each range.
« Last Edit: August 09, 2015, 03:21:25 am by retiredcaps »
Offline Lightages
  • Canadian po
« Last Edit: August 09, 2015, 03:41:07 am by Lightages »
Offline retiredcaps
« Reply #8 on: August 09, 2015, 03:28:55 am »
1. 194.5 V
2. 195.1 V
3. 199.8 V
4. 194.0 V
Wow, the UT58C reading is way off compared to the other 3.  Since the UT58C is a 2,000 count meter, this reading is close to full scale.
« Reply #9 on: August 09, 2015, 12:06:12 pm »
Maybe that is because it was not a perfect sine wave? I used a transformer on mains voltage for that measurement.
« Reply #10 on: August 09, 2015, 12:59:44 pm »
I repeated the measurement with 198 V and 130 V. All multimeters were connected the same way as yesterday. It looks much better now.
I think the deviation yesterday could be of a bad sine wave or maybe the voltage changed and the update rate of the UT58C is to slow?
Today's results:
AC Voltage:
1. 198.2 V
2. 198.8 V
3. 198.7 V
4. 197.7 V
1. 130.4 V
2. 130.9 V
3. 130.9 V
4. 130.0 V
The scope measures 131 V. But it is definitely not a perfect sine wave as you can see here:
« Last Edit: August 09, 2015, 01:06:19 pm by cyberfuzzy »
Offline joeqsmith
« Reply #11 on: August 10, 2015, 03:28:41 am »
No problem.  Glad to help.   Let me know if you figure out what the other pots are for.
Offline giorgiovi
« Reply #12 on: July 12, 2018, 02:28:31 pm »
do you have procedure calibration for MS8229?
Thanks
Offline giorgiovi
« Reply #13 on: July 12, 2018, 02:42:22 pm »
