Are both of the 61As scaled in the same way? It looks like the top 61A hasn't been auto-scaled.
Top one was auto-scaled
It is confirmed by the upper, lower values on the scale to the left
Upper (max) value: 5.000 5 V
Lower (min) value: 5.000 4 V
On the old firmware there was no more resolution in between
Adding the extra digit added the extra resolution on the vertical axis.
The new FW v2.08 shows now on the left of the trend chart:
Upper (max) value: 5.000 48 V
Lower (min) value: 5.000 47 V
This is really great and something I did not expect from a firmware upgrade
It seems like we are now having the following resolution on the trend chart:
34461A, FW1.10 +/- 100 uV Resolution
34461A, FW2.08 +/- 10 uV Resolution
34470A, FW2.08 +/- 1 uV Resolution