For those that did a warmup test of the Keysight meters, was the test from a cold start (plug it in), or from a warm start (in standby for some time and then turned on)?
Does the 7510 have a standby mode like the Keysight meters?
The Keysight meters (DMM) all do NOT have a standby mode!
The new ones do have a SW switch only, instead of a mechanical mains switch.
So even if the StandBy LED is lit, they are completely cold inside.
The initial drift in the first about 1/2 - 2h (depending on your stability criterion) is caused mainly by warming up of the resistor networks, but NOT of the voltage references.
The ovens of both the LM399H, and also the LTZ1000A warm up within 15 .. 30sec, and the output of the (more complex) LTZ1000 circuit is stable to < 1ppm within about 10 min.
See also my latest contribution in the 'Ultra Precision Reference LTZ1000' thread.
Frank
PS: I attached a warm-up curve of an LTZ1000 module, running at about 50°C.
The output really settles within less 10 minuntes, so that should not be the problem in the 34470A, even if this reference is cooked at 90°C.