P9 of the datasheet does under Max. Input voltage
https://int.siglent.com/u_file/document/SDS5000X_Datasheet_EN02B.pdf
Thanks, this scope must not be menat for use with SMPS, as only giving the voltage rating for up to 10khz, is totally no good for SMPS work....just thinking of eg flyback drain voltages etc etc etc....
Also, i assume they mean its 400vpk when you use a 1:1 probe?
Presumably if you use a 10:1 probe then its 4000Vpk?
I just connected this scope via a 10:1 probe to a node which was pulsing up to 200V for some 100ns every 3.3us.....i hope ive not damaged it?
Presumably if you use a 10:1 probe then its 4000Vpk?
Why you do these things and have oscilloscope if you write this sentence.
Have you ever looked any probe specifications.
Btw, also after DC there is also voltage derate curve vs frequency.
Some examples about other oscilloscopes.
Do you think no one is good or not meant for work with SMPS circuits.
Also one question. Why connect with just plain coaxial to oscilloscope 1M input. Also I may ask: what is coaxial length and how its capasitance affect DUT.
Please explain real reasons. Or is it just for playing fun without enough experience and knowledge until smoke come come out (do you know all electrics and electronics works using smoke and when it leaks out then it do not work anymore).
Tektronix MSO54B
Maximum input voltage
1 MΩ: 300 VRMS, CAT II
derate
at 20 dB/decade from 4.5 MHz to 45 MHz;
at 14 dB/decade from 45 MHz to 450 MHz; > 450 MHz, 5.5 VRMS
R&S®RTM3004
at 1 MΩ 300 V (RMS), 400 V (Vp),
derates at 20 dB/decade to 5 V (RMS) above 250 kHz
Teledyne LeCroy HDO4000A
Maximum Input Voltage 50 Ω: 5 Vrms, 1 MΩ: 400 V max (DC + Peak AC ≤ 10 KHz)
Teledyne LeCroy Waverunner 9000
Maximum Input Voltage 50 Ω: 5 Vrms ±10 V peak; 1 MΩ: 400 V max. (DC + peak AC < 10 kHz)
Rigol DHO4000
1 MΩ CAT I 300 Vrms, 400 Vpk (DC + Vpeak)
Take example this Rigol and connect its input to example 200MHz 200Vrms. Is it ok because well inside data sheet (except there read CAT I. Think if it cover also high frequencies)
Many many oscilloscopes do not have any specs about input max V derating vs frequency. Nothing about frequency and sure there still is limits - are there any oscilloscope user who really do not understand this. Oscilloscopes are made for educated engineers who need understand least simplest basics.One EEVblog thread for you to read...
https://www.eevblog.com/forum/testgear/rigol-ds1202z-e-input-voltage-vs-frequency-derating/Also noted you talk about short peaks, not continuous wave.
But yes, it is sad oscilloscopes inputs are not better specified and a good defense is not that others do the same.
Normally in this kind of test we use probes, not direct coaxial and even more weighted
not without any impedance matching.