I came very close to buying the AD2 a few days ago after some extensive research and comparison. In the end I did not buy it for two reasons:
One, it has very limited sample size/storage;
The second was that I got tired of having to filter the technical details that were buried in the PR department's fluff.
I do think it is a great device for the stated purpose, training/education; but am not quite so sure that it is not a little weak as a shop tool for serious analysis and troubleshooting.
I think that if they intended that use, they would have increased the memory depth in this revised model.
I read a comment somewhere on EEV Blog that National Instruments would not want this to compete with their "real" shop equipment. There may be more than a little truth in that comment.
Just a comment re: the differential inputs. They are differential with a but. Depending on power source (USB or wall wart) and other external connections on the pc, one must be careful with your ground connections.
Here is a direct quote from the reference manual that is worth reading. I almost get the idea that they do not want you to read it as it can take several readings.
A clear schematic showing these grounds would have been a lot more useful!!!
"Important Note: Unlike traditional inexpensive scopes, the Analog Discovery 2 inputs are fully differential. However, a GND connection to the circuit under test is needed to provide a stable common mode voltage. The Analog Discovery 2 GND reference is connected to the USB GND. Depending on the PC powering scheme, and other PC connections (Ethernet, audio, etc. – which might also be grounded) the Analog Discovery 2 GND reference might be connected to the whole GND system and ultimately to the power network protection (earth ground). The circuit under test might also be connected to earth or possibly floating. For safety reasons, it is the user’s responsibility to understand the powering and grounding scheme and make sure that there is a common GND reference between the Analog Discovery 2 and the circuit under test, and that the common mode and differential voltages do not exceed the limits shown in equation 11. Furthermore, for distortion-free measurements, the common mode and differential voltages need to fit into the linear range shown in Figs. 12 and 13. For those applications which scope GND cannot be the USB ground, a USB isolation solution, such as what is described in ADI’s CN-0160 can be used; however, this will limit things to USB full speed (12 Mbps), and will impact the update rate (screen refresh rates, not sample rates) of the Analog Discovery 2."
Graham