Author Topic: Keithley 2000  (Read 26916 times)
Offline EMUDTopic starter
Keithley 2000
« on: March 15, 2015, 09:01:44 pm »
I just try to understand the A/D converter of the K2000. It is a multi slope converter (it converts a current to a pulse length).
I would like to make some calculations about this converter and need some help.
If possible check your Reference voltage, the + 14V  and the -14V (these voltages are regulated by a double op amp and a transistor each) and post them here.
Thanks a lot for your help!
Thanks
EMUD
Re: Keithley 2000
« Reply #1 on: March 16, 2015, 12:14:50 pm »
Hello,
I have attached a block diagram of the A/D conversion unit. I think that this unit measures the voltages selected by U163
with comparism to the 4 voltages (0 volts (GND), 0,09811V, 1,0320V, 6,95V).
There are some A/D related tests:
Test 100.1 checks the charge - discharge circuit of the integrator formed by U137 and the components around. Measurement input is GND.
Test 100.2 checks the differences between some measurement results.
Test 101.1 does the same as 100.1 , but stores the result.
Test 101.2 does the same as 101.1 , but measurement input is Vref (6,95V nom.)
Test 101.3 checks the difference between 101.1 and 101.2 .
Best regards
EMUD
Offline TiN
  • Country: ua
    • xDevs.com
Re: Keithley 2000
« Reply #2 on: March 16, 2015, 12:21:38 pm »
So what exactly you want us to do? I have broken 2000 which I can measure for you.
Just +14V/-14V output from opamps?
Re: Keithley 2000
« Reply #3 on: March 16, 2015, 12:50:34 pm »
Hello TiN,
do not measure the +14/-14 at the op, since it is a simplified diagram, there are 2 transistors in the real world to
boost the current.
The easiest way to contact this is the resistornetwork TF-245 on position R271.
Please measure with respect to analog ground pin 1 (should be +14V) and pin 3 (should be -14V) and the
reference voltage at R315 to ground (should be +6,95V).
These voltages could be measured with the K2000 itself, you only need the positive input and the volts range.
Thank you in advance!
EMUD
P.S. The schematic and the repair manual are downloadable free in the i-net ! If I can help you with your 2000 ...
« Last Edit: March 16, 2015, 01:17:07 pm by EMUD »
Offline TiN
  • Country: ua
    • xDevs.com
Re: Keithley 2000
« Reply #4 on: March 16, 2015, 01:52:22 pm »
We know, but schematics is reversed by bbs.38hot.net member, not from vendor :)
Now, why it was important? :)
Offline ElektroQuark
    • ElektroQuark
Re: Keithley 2000
« Reply #5 on: March 16, 2015, 01:57:50 pm »
Hi EMUD,
I'm trying to repair a K2000.
Here is the thread: https://www.eevblog.com/forum/testgear/restoring-and-upgrading-rom-of-a-keithley-2000/
Any help/ideas are appreciated. Thank you.
Re: Keithley 2000
« Reply #6 on: March 16, 2015, 02:15:48 pm »
Hi TiN,
thanks for the values! I need this values to verify the values in the schematic. So, if I could collect
the values of some K2000 units we will see the tolerances and make it easier to hunt for problems.
I have seen that the +14/-14 should be double the height than the Vref. In the reversed diagram
there are some voltages given wich could lead to the wrong results in hunting for errors.
Your results : Vref 6,8470373    -V14  13,6860304   +V14  13,6983588
In the pics you have posted I see that you are quite often repairing Keithleys. Many of them use
the resistor network TF-245 . What is your opinion, does it fail often?
Thanks a lot! I will post the results here.
EMUD
« Last Edit: March 16, 2015, 03:37:05 pm by EMUD »
Offline TiN
  • Country: ua
    • xDevs.com
Re: Keithley 2000
« Reply #7 on: March 16, 2015, 04:34:00 pm »
Not so many, I mostly have 2001, which does not have TF-245 network. 2400 and 2510 have it, since they leverage almost same ADC block as 2000.
I got 2000 while ago, but had no much progress on it, so far. I saw one failed unit with TF-245 on chineese 38hot, but that's it.
I had thread about my 2000 here.
« Last Edit: March 17, 2015, 03:10:08 pm by TiN »
Offline z01z
Re: Keithley 2000
« Reply #8 on: March 17, 2015, 07:18:05 am »
Another TF-245 failure is described here. There's also info on what's inside.
TiN, you'll have to append a backslash to the topic link to make it work.
Re: Keithley 2000
« Reply #9 on: March 17, 2015, 07:36:44 am »
Hello z01z,
yes, I have read this before and put the reversed schematic and the values of the costas TF-245 together in my block diagram.
I have calculated some of the values and they show that the costas TF245 schematic and the values are ok.
Now, if we have the values of the resistors inside the network and its schematic, we only need the tolerance to test it. Calculating with the values I have got from three K2000 it has to be a high precision network, I think with 0.1 or 0.05% . So I could state that my TF-245 is gone, one of the 10K resistors has 10.5K ....
The network is available at Keithley service here in Germany for about 16€ + VAT + p&p . I have ordered and  looking forward to get it (about 4 weeks).
Many thanks to all for the information
EMUD
Offline klaus11
Re: Keithley 2000
« Reply #10 on: March 17, 2015, 07:43:56 am »
Hello z01z,
EMUD
Did you also sell the transistor snj132 in Keithley Germany?
HP3458A, HP3245a, Keithley 2000, Fluke 87V, Rigol DP832, TEK TDS5052B, HP33120A
Re: Keithley 2000
« Reply #11 on: March 17, 2015, 08:07:03 am »
Hello Klaus11,
I have sent you a PN with my contact to Keithley Germany.
EMUD
Offline klaus11
Re: Keithley 2000
« Reply #12 on: March 17, 2015, 08:44:01 am »
Hello TiN,
boost the current.
EMUD
TF245 Pin 1 +14,03 Pin 2 -14,02 to Pin 6 AGND   :-+
Re: Keithley 2000
« Reply #13 on: March 17, 2015, 08:59:43 am »
Hi Klaus11,
I have calculated your Vref :  7.015V  Is this ok ?
Thanks
EMUD
Offline picburner
Re: Keithley 2000
« Reply #14 on: March 17, 2015, 02:13:39 pm »
Hi Emud,
The resistor network TF-245 appears to have a failure rate of 43.2%.
see http://bbs.38hot.net/thread-94614-2-1.html (better to use google translate...)
I have a Keithley mod. 2306 (2 TF-245), one mod. 2010 and two mod. 2000.
The 2306 has the two TF-245 perfects as well as one of the two mod. 2000.
In the 2010 the TF-245 has one resistor open and the other 2000 has 4 resistor out of tolerance.
This seems confirm the statistic.
Could you give me also your contact at Keithley Germany?
I need 2 TF-245, thanks.
Regards
Giampiero
Re: Keithley 2000
« Reply #15 on: March 17, 2015, 03:15:21 pm »
Hi Picburner,
what do you think, are there special resistors affected or is it random? (poor quality??). Have you any information about the tolerance (I think min. 0.1%)?
What is your reference value to find the resistors faulty?
Thanks a lot
EMUD
Offline picburner
Re: Keithley 2000
« Reply #16 on: March 17, 2015, 04:11:11 pm »
Hi Emud,
I don't know the exact tolerance of the 12 resistors in the TF-245 but considering the type of instrument I agree with you for a 0.1%.
In my case I found a 20K resistor become fully open, 90K become 95K, 30,176K -> 31,365K, 1K -> 1,15K etc.
I found the internal TF-245 schematic on bbs.38hot.net.
Best regards.
Giampiero
Re: Keithley 2000
« Reply #17 on: March 17, 2015, 04:20:10 pm »
Hi Picburner,
that is my issue, a 10K has become 10.5K, so the reference is not correctly amplified to the 14 / -14V rails. (6.91V ref, 13,5V , - 13,2V)
But I think the values in the schematic are not correct. There is a schematic with other values in the net, I think this one is ok. I have attached
the .pdf .
Thanks
EMUD
Offline picburner
Re: Keithley 2000
« Reply #18 on: March 17, 2015, 04:32:57 pm »
Yes, the resistors arrangement is the same but some values don't agree....
What will be right?
Regards
Giampiero
Re: Keithley 2000
« Reply #19 on: March 17, 2015, 05:35:31 pm »
In my opinion this is these are the correct values, I think they are measured on a new TF-245, but I am not sure.
Emud
Re: Keithley 2000
« Reply #20 on: March 19, 2015, 04:49:55 am »
Here are the result of my TF-245 network, desoldered :
Pin  1 - 16   1.0013K
Pin 16 - 2    9.037K
Pin   2 - 15  10.467K
Pin  15 - 3    20.07K
Pin    3 - 14  15.04K
Pin  14 - 4    30.88K
Pin    4 - 13  30.89K
Pin    5 - 11   30.42K
Pin    6 - 9     1.0006K
Pin    7 - 9    9.418K
Pin    7 - 8     92.24K
Pin  10 - 12   27.56K
I think its gone ...
Now I am waiting for the new network, I will post the values of the new one (in about 4 weeks ...).
EMUD
    • ElektroQuark
Re: Keithley 2000
« Reply #21 on: March 19, 2015, 05:48:26 am »
Those are the values of mine, for another reference:
PIN        OHMS           PIN
1           999.58          16
2           8997.6          16
2           9998.4          15
3           19K994         15
3           14K995         14
4           30K171         14
4           29K698         13
5           30K163         11
6           999.65          9
7           8997.5          9
7           89K974         8
10         27K412         12
Offline picburner
Re: Keithley 2000
« Reply #22 on: March 19, 2015, 07:55:46 am »
Hi EdoNork,
yours measured values seem to fall within the 0.05% tolerance according to
the internal schematic posted by me earlier.
Stressing it as little as possible with the heat because it seems perfect.
The Emud's TF-245 is gone but not so worse than 2 of mine.
Regards
Giampiero
Offline TiN
  • Country: ua
    • xDevs.com
Re: Keithley 2000
« Reply #23 on: March 19, 2015, 04:20:32 pm »
Measured my 2000's TF-245.
Result: http://dev.xdevs.com/issues/1178
Matches schematics nicely.
Offline picburner
Re: Keithley 2000
« Reply #24 on: March 20, 2015, 07:03:27 am »
The argument TF-245 seems to excite a lot so I put some photos here.
the first photo shows what's inside TF-245,
the second and third photo shows the circuit that replaces the TF-245
in the new DMM pcb revision (2000 and 2015).
The fourth photo shows the pcb's layout of the old mod. 2000.
I found this on the forum bbs.38hot.net always, as seen from the watermark on the pictures ....
The following users thanked this post: ZhuraYuk