There is also a 197 Military version. Which is an uplift on the standard 197 model.
Added design/features:
- Additional shielding over the processor
- Different (shielded?) cable going to the display/panel.
- Torroids on the power input and on the meter inputs. Probably to reduce noise.
- Detachable AC power cord. This makes more sense for battery operation.
- Typically has battery/charger.
- It also has a 10A fuse on the high amp input. The standard 197/197A aren’t fused for the 10A input!!!
- Like the 197A it has 4 references in dB mode including 600 and 50 ohm. The 197 only has 600 ohm.
rastro
Edit: Just remembered the model: AN/USM-486A; http://www.liberatedmanuals.com/TM-11-6625-3055-24P.pdf
So basically, the military 197 is halfway between a 197 and 197A, plus the added fuse and CPU shield, toroid on the inputs, and with silkscreen on the analog section shielding, it looks like. (The standard 197A has a toroid on the AC jack, and the removable cord.)
Did you look at that manual? No way is it the right one; nothing about its construction matches a 197/197A, nor does it match the pictures you posted. It looks like it’s probably a Fluke design. (Looks like the Keithley 197-M is AN/USM-486A; that manual is for AN/USM-486U
Agreed, that's just not a 197. Looks like a Fluke to me, too.
Also, it looks like the transformer has a shield around it too (neither of my 197s have that - I think the 197A does).
I am still having a ton of trouble repairing my 197s. Some of the circuits seem very complicated to me. Right now one of my units is reporting OL in every range, on every function. The other intermittently reports OL in between incorrect measurements. This 2nd one (OL then measure then measure then OL) seems to to have problems with the -9V rail intermittently. It's like the negative supply isn't sufficiently loaded or something. I really don't understand how the negative supply is created/regulated.
For noobs like myself, though, it turns out that two other models rely on essentially exactly the same signal path (modulo the details of the range switching): the Keithley 175, the 192, and the 197 all have basically the same process after signal conditioning, all the way to the microprocessor.
Now if only I could get an idea of what exactly is going wrong with my units. I might have to ask someone to take oscilloscope screenshots of certain measurements. I have so many measurements I doubt are correct, though, I almost don't know where to begin.
I actually have a battery board from the 2nd unit, but I never tried it until yesterday and it turns out it arrived dead. The CMOS inverter which provides the negative rail doesn't produce any negative voltage. Eventually, I'll have to order one of those ICs.
Unfortunately I don't yet have any low noise power power supplies (especially no split-rail or bipolar ones), so I can't bypass the AC power supply yet.