Fluke 8846A error charts
« on: July 17, 2020, 06:11:15 pm »
Today I ordered a Fluke 8846A. I prepared error charts that show abs. error as a function of range & measured value. If anyone is interested, I have similar for GW Instek GDM-8351 too.

One question: What is the maximum count the 8846A can show before switching to higher range? I saw some photos displaying 1.000604V on 1V range. At which point will it switch to 10V range? Once I get this info, I will complete the charts to show the overlap of adjacent ranges.
« Last Edit: July 17, 2020, 09:18:20 pm by Baltazar »
 

Offline commongrounder

  • Frequent Contributor
  • **
  • Posts: 372
  • Country: us
Re: Fluke 8846A error charts
« Reply #1 on: July 17, 2020, 07:15:45 pm »
I have a Tektronix DMM-4050, which is essentially the same as the 8846A.  It will over-range up to 1.2 volts on the one volt range before up-shifting.  It will down shift when the voltage settles back to 1.1 volts, so there is a bit of hysteresis there. Hope that helps.
 

Offline bdunham7

  • Super Contributor
  • ***
  • Posts: 7972
  • Country: us
Re: Fluke 8846A error charts
« Reply #2 on: July 17, 2020, 07:35:21 pm »
1.2 million counts, so 20% overrange, but it is twitchy and sensitive to noise.  You'll see the word 'Overload' a lot as it shifts ranges.  If is averaging a noisy signal, it may shift down a range sooner than you'd think and may indicate 'Overload' even at much less than the full 1.2M counts.  By 'error chart' do you mean a compilation of tolerances from the data sheet or from actual measurements of your unit?
A 3.5 digit 4.5 digit 5 digit 5.5 digit 6.5 digit 7.5 digit DMM is good enough for most people.
 

Offline BaltazarTopic starter

  • Contributor
  • Posts: 14
  • Country: 00
Re: Fluke 8846A error charts
« Reply #3 on: July 17, 2020, 09:22:45 pm »
Thanks for the info, commongrounder & bdunham7. I updated the charts, i.e. extended every range by 20%.

@bdunham7: This is a compilation of ± (% measurement + % of range) values from spec.
 